BAXTER 6201

(19)
(8)

yi huang

BAXTER 6201

February 24, 2006 09:26

Dear sir,
A BAXTER 6201 INFUSION PUMP report F_94 when it was turned on.Who can give us some advice kindly?
Thank first!
htdocument@126.com

Log inor Register

to rate and post a comment

(3)
(1)

Chuck Rowland

Re: BAXTER 6201

February 24, 2006 10:56

This error is because all of the config options have reset to zero. Check that the main battery and backup battery are good. Reset the config options.Make sure that in the first screen that you clear the alarms. Then power on the pump. 90% of the time this will clear the error.
If you have any problems, feel free to give me a call at (800) 625-3551 and I can assist you in clearing up this error.

Chuck Rowland, C.T. Medical Electronics

Log inor Register

to rate and post a comment

(10)

(7)

Donald Creedon

Re: BAXTER 6201

February 24, 2006 11:07

Call Phillip at Signature medical. Web site is www.sepems.com

Log inor Register

to rate and post a comment

(4)

Cliff Hall

Re: BAXTER 6201

February 24, 2006 12:24

The F-94 is a configuration setting code, you will most likely need to replace the main battery and reset the date and time. I would try to find a manual to help you in getting to the correct parameters to reset the date and time.