Ampronix, a Top Master Distributor for Sony Medical, provides Sales, Service & Exchanges for Sony Surgical Displays, Printers, & More. Rely on Us for Expert Support Tailored to Your Needs. Email info@ampronix.com or Call 949-273-8000 for Premier Pricing.
edward bartra
Bennett Contour error message
April 07, 2006 04:25
Bennett contour error message.
coll and Relay
dont have books, anyone know what error codes these are or any direction to go.
thanks
to rate and post a comment
Felix Donoghue
Re: Bennett Contour error message
April 08, 2006 02:53
I have a complete service manual, if you need it i can copy and send you one.
Thanks
to rate and post a comment
edward bartra
Re: Bennett Contour error message
April 08, 2006 12:22
thank you but we will be getting a service manual. we need to get the machine running to shaow to a customer and with these two errors it won't work. is anyone familiar with this machine?????
to rate and post a comment
Deon Armes
Re: Bennett Contour error message
April 09, 2006 12:46
We were a Bennett Dealer for five years (94-99)in Tampa. The fact that you have two errors from totaly different places on the machine tells me there is possibly a bad PIA chip on the CPU PCB they are 6829 chips on CPU PCB. The coll error usually says the auto collimator is not at the right size (stuck). The relay error says the backup contactor did not come in upon prep. I would swap the PIA chips on the CPU PCB and see if the errors change or you have different problems. They are in sockets so this is easy. If you need a schematic for the CPU PCB call me I can fax it to you.
Good luck!
Deon
to rate and post a comment