Blurred image when making fluoroscopy

Tags: C-Arm, PHILIPS

(31)
(2)

AnyFE

Blurred image when making fluoroscopy

January 09, 2026 05:54

Hello, dear group.
I´m servicing to a C-Arm Pulsera from Philips. When making fluorospopy get blurred images. I´ve already checked voltages on image intensifier power supply. Voltage measured at test point Vg2 is +3.15v, but according to technical specification on service manual, it should be +0.8v. Those grid voltage values make me think that II power supply is damaged, but I´m not sure about that. Could anybody help me to make a better diagnostic? Your assistance will be unvaluable.
Thanks before hand.
Regards

Log inor Register

to rate and post a comment

(265)
(50)
(28)

Wayne Horsman

re: Blurred image when making fluoroscopy

January 12, 2026 10:15

adjust g1 voltage to balance edge focus to center focus, g2 is the main focus adjustment, g3 is the size adjustment. They all interact a little, so adjusting one a lot may require a minor tweak of the other two

Log inor Register

to rate and post a comment

(9)

medmike

re: Blurred image when making fluoroscopy

January 12, 2026 10:15

Hello!
Have you tried the other zooms, too ?
Same issue ? Is it completely unsharp or only little bit ?
Usually the focus can be adjusted at II power supply, but only by smaller deviations

You can also remove the camera from II and hold it in the room ….if room is sharp seenable you can be sure that problem is not from camera side -

Log inor Register