Booting problem with DARC2 LightSpeed CT

September 09, 2015 03:23

Hi guys,

I'm having a booting issue on the GOC4 console for the LightSpeed, so everything seems booting normal until the HP host doesn't connect properly with the darc, and waits for it to boot up completely, but even if the darc booted properly, the host doesn't see it sometimes.

I've tried to verify the connection:
- ping darc (succesful)
- rsh darc (succesful)
- ping ig1 (succesful) from darc
- ping localhost (succesful) from darc
- ssh darc (succesful)
- service cliservice start
telnet localhost 623
Server: darc
Username:
Password:
Login succesful

- dpccli/ reset
ok
- dpccli/ console (at this point the DARC resets again and when its done, there is a shell window with the following script)
.........
/usr/g/bin/gre-verify-darc: built May 11 2007

ping localhost: waiting...done
ping darc: waiting..........done
ping darc echo: waiting...ssh: connect to host darc port 22: Connection refused....done

OC application version: 07MW11.10.H2_P_M4_G_Zeus
DARC application version: 07MW11.10.H2_P_M4_G_Zeus

darc:/usr/g/config/scanhardware.cfg: rsyncing...done
darc:/dev/shm/gre-raid-lock/gre-raid-exit: waiting...done
darc:/usr/g/service/log/gre-raid-exit-status.log: rsyncing...done
darc:/dev/shm/gre-raid-lock/gre-raid-logmsg: removing...done
darc:/dev/shm/gre-raid-degrated:checking...not found (RAID is not degraded)
darc:/proc/mounts:/raw_data...bash: line 1: 8197 Terminado
attention -nb -ontop -rn GreVerifyDarcWaitingForStarup 2/dev/null /dev/null mount is unknown
gre-raid: failure - not enough hard drives found (exit status: 16)
/usr/g/bin/gre-veryfy-darc: failure -scan disk array is non-recoverable (exit status: 22)

PS. The DARC by itself runs good until See you on the darc side of the moon.

Any help would be appreciated, thanks in advance.

1 person liked this
Log inor Register

to rate and post a comment

(423)
(40)
(5)

Juan de la Cruz Martinez

re: Booting problem with DARC2 LightSpeed CT

September 09, 2015 06:40

gre-raid: failure - not enough hard drives found (exit status: 16)
If I was you I reload the software on the DARC, you also might have on the drives on the DARC bad, see the warning above

2 people liked this
Log inor Register

to rate and post a comment

(140)
(35)

FranciscoG

re: re: Booting problem with DARC2 LightSpeed CT

September 12, 2015 08:42

Thanks I did this but the commands didn't work, so I tried reformatting the database, and it failed. So I will be replacing both disks I've got.

1 person liked this
Log inor Register

to rate and post a comment

(62)
(22)

Youbissi Pascal

re: Booting problem with DARC2 LightSpeed CT

September 10, 2015 08:53

Firt of all you need to reload software to the DARC2. If after that the same problem happens, It seems that one of you have a hard problem with your RAID HDD, replace IT.

Regards

1 person liked this
Log inor Register

to rate and post a comment

(140)
(35)

FranciscoG

re: re: Booting problem with DARC2 LightSpeed CT

September 12, 2015 08:42

Thanks, I will be replacing the HDD.

Log inor Register

to rate and post a comment

(14)
(7)

Syl Harris

re: Booting problem with DARC2 LightSpeed CT

September 10, 2015 10:16

sounds like at least 2 of the raid hard drives has failed. The raid drives should indicate failure with a red LED vs. a green LED, indicating an okay status.

1 person liked this
Log inor Register

to rate and post a comment

(140)
(35)

FranciscoG

re: re: Booting problem with DARC2 LightSpeed CT

September 12, 2015 08:42

Thanks, I will be replacing the HDD.

Log inor Register

to rate and post a comment

(30)
(9)

Dean Kenney

re: Booting problem with DARC2 LightSpeed CT

September 11, 2015 06:10

The easiest way to check out the two drives is apps shutdown reconfig check the regeneration data base and read the scrolling scripts. You will see raid success or failure.

1 person liked this
Log inor Register

to rate and post a comment

(140)
(35)

FranciscoG

re: re: Booting problem with DARC2 LightSpeed CT

September 12, 2015 08:42

Thanks man, I did this already and got a failure... so I'm replacing the HDD.

Log inor Register

to rate and post a comment

stats
DOTmed text ad

MRI & CT PARTS USED-TESTED, REFURBISHED. Call (512) 477-1500

Extensive inventory of Philips & Siemens MRI & CT used-tested, refurbished parts. Knowledgeable Sales and Inventory Team. 90-Day Warranty. Bilingual Support. Competitive Pricing. Technical Support. ISO Certified. parts@keimedicalimaging.com

DOTmed Certified stats

Upcoming Auction Events

Cbrx2sales
TIMED Ending Soon
Jun 05, 2026 (157 lots)