Brightspeed 10V reference error

(27)
(7)

John7475

Brightspeed 10V reference error

December 18, 2015 08:56

I need your help on this problem of GE brightspee8.. 
The machine will pass on ddc scan without rotor on.. but when the rotor on selected, it gives an error.

The error was like a DC bus error (70-0501). But the measured value on hv inverter was 651vdc(on pdu). Same on JEDI and hemit tank.
The other error that makes me confused is the 10V reference error (80-1419). There is no explanation on manual error code..

I do not know where to check or verify regarding this 10V reference error.

tanx for reply..

Log inor Register

to rate and post a comment

(14)
(7)

Syl Harris

re: Brightspeed 10V reference error

December 18, 2015 06:40

70-0501, Potential cause: 1. mains proglem (too low or too high)or 2. One phase missing at generator input Recommended action:(CHECK FUSES IN PDU OR ON GENERATOR)
If hard fault, check fuses in PDU (mains input and HVDC fuses, gantry base and HV inverter.
-If fuse open, hceck for load shorts before replacing.
-Check mains line, 3 phases incoming voltage.
-Verify line impedance if mains is low.
-Check the PDU main fuses status

80-0149, Possible cause: PPC kV control board
Recommended action: Replace PPC kV control board.

Log inor Register

to rate and post a comment

(27)
(7)

John7475

re: re: Brightspeed 10V reference error

December 19, 2015 09:58

hello Mr. Syl, thank you for reply.. yes i did check the 3phase input on the diode for hv inverter(pdu), the 3 phase suply was 401,404,402vac. the output is 651vdc.. error 70-0501 was now gone.. but when i do DDC scan with rotor on, always error 80-0149 was appeared..

i will try to replace the PPC kv control board and update you later.. tanx

Log inor Register

to rate and post a comment

(183)
(13)
(2)

Borhan Kalash

re: re: Brightspeed 10V reference error

December 19, 2015 06:24

I do have experiance with brightspeed just hispeed and lightspeed
I will check the power supply that supply the kv control board for such trouble