Over 150 Missouri Auctions End Today - Bid Now

Brightspeed 16 error 26014033

(2)

Pablo Gutierrez

Brightspeed 16 error 26014033

March 29, 2018 10:50

Hello
I need help I am working with a CT scan brightspeed 16, All in One console, the problem is that the gantry is not available to do studies, the gantry and the table are online, I can move them, but when I want to do the scan it tells me that it is not available, by the way, i reload the software and even then it does not work.
Error: 290009035
ERROR: can not process ASIR request - process is not up
ERROR: 26014033
A plug n play device has encounters an unrecoverable error. This device will remain offline until a scan hardware reset is performed
led DS04 on TGPU its ON
led DS05 on TGPU its ON
ERROR: 245183
Did not receive reply from a request

Log inor Register

to rate and post a comment

(410)
(57)

CRAIG WAMBURA

re: Brightspeed 16 error 26014033

March 30, 2018 10:01

did you perform a flash download plus a scan hardware reset?