May 28, 2019 10:05

BrightSpeed failled to initialize gantry
Error 260140019
Some controllers(TGP, ORP, IPC, PROP, PSC) do not respond to ping
I ping tgp and orp with succes

Please help

Log inor Register

to rate and post a comment

(410)
(57)

CRAIG WAMBURA

re: Brightspeed 16 : failled to initialize hardware

May 29, 2019 10:32

Check in PDU if all circuit breakers are up.

Is PDU gantry enable on?

Green light in front of PDU next to Estop

Log inor Register

to rate and post a comment

(43)
(1)

Changsheng Liu

re: Brightspeed 16 : failled to initialize hardware

May 29, 2019 10:32

If the version before 2013
Can help

Log inor Register

to rate and post a comment

(410)
(57)

CRAIG WAMBURA

re: Brightspeed 16 : failled to initialize hardware

May 29, 2019 02:37

contact me via whatsapp +256782702243

Log inor Register

to rate and post a comment

(74)
(15)

PIEBENG APPOLIN

re: re: Brightspeed 16 : failled to initialize hardware

May 30, 2019 12:32

Thank for your reply
I allready solve problem, it was wrong connection between console intercom board and gantry TGPU