after battery replacement CT Brightspeed got this:
Host: Brightspeed Proc:scanRx Error: 200501533
File: uiflpcCommonFunctions.c Line:2272
Function: No system function reported
Sevice uif_scan_rxfail to receive a message with event code 3213 from Event Router process with status:
Timed out while attempting to receive message

Mon Oct 20 ...
Host: Brightspeed Proc:scanRx. Error: 200501547
File: uiflpcMgrScanRx.c Line: 11478
Function: Data Acquisition: OC Processing
A Communication Error has occurred in uiflpcSenRecvTubeStateRequest with seq_num = 7 and status=200001821

Mon Oct 20 ....
Host: Brightspeed Proc: scanRx Error: 200501544
File: ZeusrpGlobalActions.c Line :1839
Function: Data Acquisition: OC Processing
getTubeWarming Timings(): Failure had ocurred while performing rxif_getTubeState

any comment to solve it?

Log inor Register

to rate and post a comment

(8)

medicorpchile

re: Brightspeed Communication Problem

October 22, 2025 08:50

What battery?

Log inor Register

to rate and post a comment

(2)

FDMEnterprises

re: Brightspeed Communication Problem

January 19, 2026 04:01

When I have seen this error it is usually related the the small din 9 connector on the tube. I would check that plug. I am guessing the battery you are referring to is on the Jedi generator and you are trying to do generator cals and got this error. I believe there is a connector from the generator to the tube that may have been bumped during the battery change or the din 9 that goes to the tube has been disconnected.

Log inor Register

to rate and post a comment

stats
DOTmed text ad

PHILIPS MX8000 & Brilliance Systems & Spare Parts In Stock Now!

HIGH QUALITY fully tested & refurbished systems to manufacturer specifications like new. Installation & service support worldwide. Parts & tubes in stock with warranty. Sales@Elsmed.com

DOTmed Certified stats

Upcoming Auction Events

Medical Equipment Superstore
TIMED
Jul 14, 2026 (502 lots)