Good evening friends, I have a problem in a BrightSpeed it presents the error 0313H 787 ILR max current detected
0403H 1027 mA out of range
0314H 788 ILR current timeout
The console also warns a current resonant inverter tube was changed recently worked 2 days and gave this error in test KV noload spit the max error someone has gone through the same thing? I can not even calibrate the machine because the system aborts the error may be in JEDI? or tank?
I would say Tank also. If the no load KV test fails with the tube disconnected from the circuit, it's most probable the tank. But you should do the other KV tests to be sure.
HIGH QUALITY fully tested & refurbished systems to manufacturer specifications like new. Installation & service support worldwide. Parts & tubes in stock with warranty. Sales@Elsmed.com
Caio Pgri
BrightSpeed Error 60 0313H
November 01, 2012 05:39
Good evening friends, I have a problem in a BrightSpeed it presents the error 0313H 787 ILR max current detected
0403H 1027 mA out of range
0314H 788 ILR current timeout
The console also warns a current resonant inverter tube was changed recently worked 2 days and gave this error in test KV noload spit the max error someone has gone through the same thing? I can not even calibrate the machine because the system aborts the error may be in JEDI? or tank?
to rate and post a comment
Gustavo Gnecco
re: BrightSpeed Error 60 0313H
November 02, 2012 09:51
Hi Cassio,
If you have only the 0313H, our experience would say the Jedi tank, but you have the 0314H, so probably is the KV Control
The 0403H probably means you had a spit, but it will not stop the exposure, just log
Have you tried the others generator test ??
Run this test to confirm if it is KV Control Board
-Inverter in Short Circuit Diagnostic
-Inverter Gate Command Diagnostic
Please let me know the results
Regards
Gustavo
www.emsts.com.br
PS: Acabei de ver que vc é do Brasil, se precisar de assistência entre em contato
Abraços.
to rate and post a comment
Artemio Santacruz
re: BrightSpeed Error 60 0313H
November 02, 2012 09:51
Did You tried to do it manually calibration? you can also do it manually ...
to rate and post a comment
Mario Wug
re: BrightSpeed Error 60 0313H
November 03, 2012 11:00
I would say Tank also. If the no load KV test fails with the tube disconnected from the circuit, it's most probable the tank. But you should do the other KV tests to be sure.
to rate and post a comment
Haroon Rashid
re: BrightSpeed Error 60 0313H
November 03, 2012 11:00
Hi Cassio,
I think the problem is in your Tank because i face the same problem in HiSpeed and i detect the tank faulty run generator test
-Inverter Gate Command test
-Inverter in Short Circuit test
-No Load KV test
If first two tests pass then do the NO LOAD KV test with different techniques and different exposure time.
Regards,
Haroon Rashid
to rate and post a comment