BrightSpeed Error 60 0313H

(20)
(5)

Caio Pgri

BrightSpeed Error 60 0313H

November 01, 2012 05:39

Good evening friends, I have a problem in a BrightSpeed ​​it presents the error 0313H 787 ILR max current detected
0403H 1027 mA out of range
0314H 788 ILR current timeout

The console also warns a current resonant inverter tube was changed recently worked 2 days and gave this error in test KV noload spit the max error someone has gone through the same thing? I can not even calibrate the machine because the system aborts the error may be in JEDI? or tank?

Log inor Register

to rate and post a comment

(51)
(24)
(3)

Gustavo Gnecco

re: BrightSpeed Error 60 0313H

November 02, 2012 09:51

Hi Cassio,

If you have only the 0313H, our experience would say the Jedi tank, but you have the 0314H, so probably is the KV Control

The 0403H probably means you had a spit, but it will not stop the exposure, just log

Have you tried the others generator test ??

Run this test to confirm if it is KV Control Board
-Inverter in Short Circuit Diagnostic
-Inverter Gate Command Diagnostic

Please let me know the results

Regards

Gustavo
www.emsts.com.br

PS: Acabei de ver que vc é do Brasil, se precisar de assistência entre em contato
Abraços.

2 people liked this
Log inor Register

to rate and post a comment

Artemio Santacruz

re: BrightSpeed Error 60 0313H

November 02, 2012 09:51

Did You tried to do it manually calibration? you can also do it manually ...

1 person liked this
Log inor Register

to rate and post a comment

Mario Wug

re: BrightSpeed Error 60 0313H

November 03, 2012 11:00

I would say Tank also. If the no load KV test fails with the tube disconnected from the circuit, it's most probable the tank. But you should do the other KV tests to be sure.

1 person liked this
Log inor Register

to rate and post a comment

(66)
(8)

Haroon Rashid

re: BrightSpeed Error 60 0313H

November 03, 2012 11:00

Hi Cassio,

I think the problem is in your Tank because i face the same problem in HiSpeed and i detect the tank faulty run generator test

-Inverter Gate Command test
-Inverter in Short Circuit test
-No Load KV test

If first two tests pass then do the NO LOAD KV test with different techniques and different exposure time.

Regards,
Haroon Rashid

1 person liked this
Log inor Register

to rate and post a comment