Over 150 Missouri Auctions End Tomorrow 06/05 - Bid Now

Brightspeed HV ERROR

Tags: CT

Posted by an Anonymous User

Brightspeed HV ERROR

February 23, 2017 10:24

I got throubles on GE Brightspeed on XRay ON, using 100MA or more
60-313 Inverter Max IRL or
69-318 kV did not reach 75 %
I tested inverter Short circuit OK, Rotor and Heater OK too
Any suggest about solve this problem
Thank you

Log inor Register

to rate and post a comment

Mateo Barloqui

re: Brightspeed HV ERROR

February 24, 2017 10:37

Hi, how are You?
According to what you say your problem in the HV tank. You would confirm it testing No Load kV!
I hope that this helps you! Regards

Log inor Register

to rate and post a comment

(9)
(4)
(6)

Brent Hull

re: Brightspeed HV ERROR

February 26, 2017 02:35

I have seen this error on a BrightSpeed 4. The fix was to replace the heater board.

1 person liked this
Log inor Register

to rate and post a comment

(9)
(4)
(6)

Brent Hull

re: Brightspeed HV ERROR

February 26, 2017 02:36

I have had this problem. Replace the Heater board.

1 person liked this
Log inor Register

to rate and post a comment

(34)
(8)

Ruben Cadima

re: Brightspeed HV ERROR

February 26, 2017 02:36

My sugestión. Try NOLOAD KV TEST with TUBE cable phisically desconnected from hv thank to test this ítem individualy, then verifie the imoedance of hv cable pin to pin betwen both sides.
According to the results the problem could be HV TANK, cable or TUBE.
Regards

Log inor Register

to rate and post a comment

(410)
(57)

CRAIG WAMBURA

re: Brightspeed HV ERROR

February 27, 2017 10:43

perform KV Loop Test and KV MA meter checks to isolate the HV cables and HV Tank/Inverter.

1 person liked this
Log inor Register

to rate and post a comment

stats
DOTmed text ad

JOHNSON & JOHNSON MONARCH QUEST - For Sale $299k or Lease $6k Per Month, 48 Months

Lightly used system available direct from the end user customer. MONARCH QUEST Featuring our most advanced, data-driven navigation software with AI-Powered algorithms & advanced imaging integration with the GE HealthCare OEC 3D C-arm.

stats

Upcoming Auction Events

Cbrx2sales
TIMED
Jun 05, 2026 (157 lots)