I have problem with vertical movement of the table, it does not move up or down. Only when I press service key in the left side of the gantry the table may move up and down. The gantry can't be tilted or rotated. I appreciate If any body can help me with this or send me the service manual.
Hi Dina!
Check under the table below the integrity of the parts: a rubber ring at the beginning of the table on the motor, and a rope of a vertical potentiometer at the end of the table in front of the gantry
I forget to tell that this problem happened when I switch to service key of the gantry and then use the up & down switches at the bottom of the table. May be some thing happened to one of the limit switches. My question is how many limit switches do I have? Because I recognize 6, and not sure if there is any more?
HIGH QUALITY fully tested & refurbished systems to manufacturer specifications like new. Installation & service support worldwide. Parts & tubes in stock with warranty. Sales@Elsmed.com
Dina Alnahdi
Brilliance 16
April 08, 2019 10:25
Hi every one,
I have problem with vertical movement of the table, it does not move up or down. Only when I press service key in the left side of the gantry the table may move up and down. The gantry can't be tilted or rotated. I appreciate If any body can help me with this or send me the service manual.
to rate and post a comment
Joseph Jenkins
re: Brilliance 16
April 09, 2019 10:20
email me at iiltd@hotmail.com
to rate and post a comment
Aleksander Proof
re: Brilliance 16
April 09, 2019 10:20
Hi Dina!
Check under the table below the integrity of the parts: a rubber ring at the beginning of the table on the motor, and a rope of a vertical potentiometer at the end of the table in front of the gantry
to rate and post a comment
raj venugopal
re: Brilliance 16
April 09, 2019 10:20
check the ACS unit
to rate and post a comment
raj venugopal
re: re: Brilliance 16
April 10, 2019 08:57
hi send the 7segment display of ACS unit in placed in table top than service mode switches on manuvel up and down is working....?
to rate and post a comment
Indarpersad Singh
re: Brilliance 16
April 09, 2019 10:21
I am not an expert, but you can check Patient Table Board D702. Controls Vertical and also interface between PHS and Gantry.
to rate and post a comment
raj venugopal
re: re: Brilliance 16
April 10, 2019 08:57
you sent the 7 segment led display of ACS control in table
to rate and post a comment
Dina Alnahdi
re: re: re: Brilliance 16
April 10, 2019 04:57
7 segment shows number "3"
to rate and post a comment
raj venugopal
re: re: re: re: Brilliance 16
April 11, 2019 11:52
sir send the error detail my mail id rajvenugopal@greenmedichealthcare.com
send the your mobile no
to rate and post a comment
Dina Alnahdi
re: re: re: Brilliance 16
April 10, 2019 04:58
The 7 segment led display shows number "3"
to rate and post a comment
raj venugopal
re: re: re: re: Brilliance 16
April 11, 2019 11:52
sir you send the error detail and your mobile no my mail id rajvenugopal@gmail.com
to rate and post a comment
Dina Alnahdi
re: re: re: re: Brilliance 16
April 11, 2019 11:52
I forget to tell that this problem happened when I switch to service key of the gantry and then use the up & down switches at the bottom of the table. May be some thing happened to one of the limit switches. My question is how many limit switches do I have? Because I recognize 6, and not sure if there is any more?
to rate and post a comment