Brilliance 64-Gantry cannot be initialized

Tags: CT, PHILIPS

(5)
(3)

MKuatly

Brilliance 64-Gantry cannot be initialized

May 20, 2019 09:27

Hi,

I have problem in Brilliance 64, system is giving this message " Gantry cannot initialize, please change scan parameters or restart the Gantry. If problem persist, please contact Service" directory after registering new patient and start exam.
---------
I checked ErrorLogs and found the below errors:
1- S_MASTER_SEQUENCER_PREP_STARTER_SPIN_ERROR.
2- S_ROTOR_STOP_ANODE_NO_MORE_SPINNING.

- I tried to do Manual Scan but message "Tube is over heated. Please wait for a while" will appear while the tube heat is 0%.
- With No Load test and Performance Test and after pressing the system will try to load series for a while then timeout error will appear.

I replaced APM but same error.

Kindly, your support and help

regards,

1 person liked this
Log inor Register

to rate and post a comment

(39)
(3)

GNES

re: Brilliance 64-Gantry cannot be initialized

May 21, 2019 09:15

hello write me to gnesker@gmail.com

Log inor Register

to rate and post a comment

(39)
(3)

GNES

re: Brilliance 64-Gantry cannot be initialized

May 21, 2019 09:16

hello write me to gnesker@gmail.com

Log inor Register

to rate and post a comment

(423)
(40)
(5)

Juan de la Cruz Martinez

re: Brilliance 64-Gantry cannot be initialized

May 21, 2019 09:18

It seems like you have a rotor problem, either the tube or associated circuits

Log inor Register

to rate and post a comment

(2)

Sebastian Gwozdz

re: Brilliance 64-Gantry cannot be initialized

May 23, 2019 09:37

could be rhost problem or tube by self. Write me at sebastian.robert.gwozdz@gmail.com