Brilliance 64-Gantry cannot be initialized

May 20, 2019 09:22

Hi,

I have brilliance 64 which is giving message "Gantry cannot initialize, Please change scan parameters or restart the Gantry" directly after registering the patient and start the exam.

- I did manual scan but after executing series, the system will show message "Tube is over heated. Please wait for a while"
- I did performance check and no load check but after executing series, the system will show "Timeout waiting for series Ready state"

- I checked the error-logs and found this below two error is appearing:
1- S_ROTOR_STOP_ANODE_NO_MORE_SPINNING
2-S_MASTER_SEQUENCER_PREP_STARTER_SPIN_ERROR

Kindly, your advice: mkuatly@gmail.com

2 people liked this
Log inor Register

to rate and post a comment

(29)
(4)
(4)

William Dixon

re: Brilliance 64-Gantry cannot be initialized

May 21, 2019 09:51

Sounds like more toward the RHOST on the CPM instead of the APM side with the tube starter.

Log inor Register

to rate and post a comment

(29)
(4)
(4)

William Dixon

re: Brilliance 64-Gantry cannot be initialized

May 21, 2019 09:51

But don't forget the DMC -- try reflashing / download both Rhost and DMC in CPM utilities

Log inor Register

to rate and post a comment

(107)
(11)
(3)

Gustavo Volpi

re: Brilliance 64-Gantry cannot be initialized

May 26, 2019 12:04

We can help, contact us! Regards

Log inor Register

to rate and post a comment