Burdick Quest 4.0

October 17, 2012 09:33

Burdick quest does not print interpretation in resting mode, or any mode, as burdick says it should. Any ideas what the problem is. Tried 3 different rev software and 3 different rev boards. After the patient data on the top of the page there is nothing, no complexes no interpretation. Everything else prints fine as well as the system function perfect.

Log inor Register

to rate and post a comment

(17)
(2)
(3)

Ramon Cervantes

re: Burdick Quest 4.0

October 17, 2012 11:20

the cable to the patient module is probably bad or not fully connected.look on entrance to machine and entry port on board. the unit is not picking up signals. try another patient module.

Log inor Register

to rate and post a comment

(14)
(2)
(2)

Nils Nilson

re: re: Burdick Quest 4.0

October 18, 2012 12:48

All leads are displaying perfect on the monitor and the printout, also all patient data. Missing to the right of the Patient data is Interpretation only. Client also said it was there before

Log inor Register

to rate and post a comment

(17)
(2)
(3)

Ramon Cervantes

re: re: re: Burdick Quest 4.0

October 18, 2012 12:39

YOU HAVE A DIRTY PRINT HEAD,TAKE ALCHOL PREP AND SWIPE COMPLETE PRTINT HEAD AND ALLOW TO DRY. DO THIS WHILE MACHINE IS OFF,GOOD LUCK

Log inor Register

to rate and post a comment

(14)
(2)
(2)

Nils Nilson

re: re: re: re: Burdick Quest 4.0

October 18, 2012 03:09

It prints patient data on the top, not dirty print head

Log inor Register

to rate and post a comment

(161)
(7)

ultrasoundman

re: Burdick Quest 4.0

November 06, 2012 11:39

I need Interpretation Option software

Log inor Register