Burdick/siemens EKG e550

September 11, 2006 05:22

I have a e550 when turned on the display just flashes for a second and then does nothing. No service manuals etc. Any ideas? Thanks

Log inor Register

to rate and post a comment

CLARENCE KNOX

Re: Burdick/siemens EKG e550

September 12, 2006 05:26

I thank that if you check the battery in your ECG e550, you will see that it is bad. Replacing it should do the trick.

Clarence Knox, BMET
bmedknox@msn.com

Log inor Register

to rate and post a comment

Ernesto Garcia

Re: Burdick/siemens EKG e550

September 14, 2006 07:57

Hi:

Check the battery's expiration date, if you change the battery I bet you solve this problem

Good Luck
Ernesto

Log inor Register

to rate and post a comment

(9)
(1)

Lee Smith

Re: Burdick/siemens EKG e550

October 26, 2006 09:44

Are you talking about the main battery or the backup? I've replaced the main lead acid and is still doing the same thing.

Thanks

Log inor Register