Abdelmuin Khatib

BV 25 , C-ARM

March 04, 2012 06:44

i have BV 25 , c-ARM machine , this machine give us error 1 , this probem from casscatt II , (powe intensifier supply ) , i removed it and clean it from silicon , the error gone , but the image not clear very cloudy .
i hope somebody to help me and tell me what the reason .

Log inor Register

to rate and post a comment

(423)
(40)
(5)

Juan de la Cruz Martinez

Re: BV 25 , C-ARM

March 05, 2012 08:56

Did you look at the image on the II output (Green) before you mounted the camera tube?
It probably is a bad supply, check the voltage of the power supply with a hight tension probe meter.

Log inor Register

to rate and post a comment

(59)
(5)
(4)

Abdelmuin Khatib

Re: BV 25 , C-ARM

March 06, 2012 12:28

the probem is how i can mussure this voltage , it is more than 40kv .
i hope to help me.

Log inor Register

to rate and post a comment

(50)
(5)

ricardo Usandivaras

Re: BV 25 , C-ARM

March 06, 2012 10:50

You need one high voltage prove,used with your digital multimeter,it give 1 mv= 1KV in trhe escale of 2v=2000mv

Log inor Register

to rate and post a comment

(16)
(2)

Steven Zhao

Re: BV 25 , C-ARM

March 05, 2012 10:21

your problem with Casscate II generator, bad focus with Image intensifier due to defective II generator , you can adjust the II generator(potentiometers inside the Casscate ii generator)if still have problem, you should change the casscate generator.

Log inor Register

to rate and post a comment

(35)
(7)
(2)

klaus boehmdorfer

Re: BV 25 , C-ARM

March 13, 2012 08:22

high voltage supply costs from philips 5000€
you should buy an other one
at all philips c-arms the high voltage supply for the intensifier is a problem.
in germany a company repairs the power supply for 3.200 €