Over 500 Total Lots Up For Auction at One Location - NJ 06/15

BV 29 Problems - Help Required

(26)
(6)
(22)

Rajagopal Geethu

BV 29 Problems - Help Required

June 02, 2008 04:42

The BV 29 which was imported from US for use here in India, gives the following error codes on being switched on:

We have used a step down transformer to get an output of 110V - 50 Hz

Error code 1 and 4

The backlight of the digital display does not get illuminated. We have checked all the power supply voltages. Except 15 volts all are coming properly - only 8 volts is coming on the 15 volts point. When the X2 connector in the Image Intensifer is removed the system loads up without any error code but we are not getting any exposure.

Is this a power supply problem in the Image Intensifer??

Any advice would be appreciated

Raj

Log inor Register

to rate and post a comment

Rafael Cruz

Re: BV 29 Problems - Help Required

June 27, 2008 08:56

Error 1 TV Alarm
Error 4 Mains control alarm

Log inor Register