bv pulsera c-arm

March 31, 2015 04:15

what is an m208 error on a bv pulsera c-arm

Log inor Register

to rate and post a comment

(72)
(18)

tim cowell

re: bv pulsera c-arm

April 01, 2015 05:19


Display code: 208
Display text: Subsystem error, system still available
Display text: It's possible to continue
Logfile text:
Unexpected X-ray switch off. Check board SHA4