bv pulsera c-arm

October 06, 2020 10:16

bv pulsera giving errors m208 and m343. Info says problem with sha4 pcb or cabling. is cabling the interconnect cable or another cable.

Log inor Register

to rate and post a comment

(82)
(15)

Herrmann

re: bv pulsera c-arm

October 07, 2020 11:13

Hi,

Have you this Errors by high KV? Maybe by lateral Position of the C-Arm?

KR Willi

Log inor Register

to rate and post a comment

(7)

rockbob

re: bv pulsera c-arm

October 12, 2020 09:50

check for fan spinning and for dust or debris blocking opto that monitors fan spinning

Log inor Register

to rate and post a comment

(72)
(18)

tim cowell

re: bv pulsera c-arm

October 14, 2020 10:03

If you cannot get a successful performance test And you are getting these errors. The most likely cause is gonna be the x-ray tube.