BV25 PROBLEM WITH THE MEMORY

(12)
(3)

Mourad Ben Salah

BV25 PROBLEM WITH THE MEMORY

January 12, 2008 01:01

Dear Sir,
i have a BV25 Carm,at first it works in good condition but after some timethe first photo that shall jump in the second screen dont appear in the second screen.the first screen still always take the new image but the second screen dont show the previous image.
we suggest that the problem is in the memory.
we dont have any documentation.
What maybe the problem?what's the defected boards (platine)?
can be the accumulator of the memory board is defected?(if there is accumulator)
we think the defect platine is in the monitor chart part
the problem appear after almost 20 minutes of work.in the beginning it works good but during the operation the problem occur.
If you have any other suggestion ,please help
I wait yr reply
Best Regards
Ben Salah

1 person liked this
Log inor Register

to rate and post a comment

saju bhaskaran

Re: BV25 PROBLEM WITH THE MEMORY

January 20, 2008 11:34

Dear Ben
under the posting where we were interacting found to be deleted from dotmed.????
Please let me know i posted a reply on further checks did u see my posting on how to go about the problem.please reply.are u base din US.we have plans to start our operations in US.
Regards.

1 person liked this
Log inor Register

to rate and post a comment

(12)
(3)

Mourad Ben Salah

Re: BV25 PROBLEM WITH THE MEMORY

January 21, 2008 10:34

Dear Mr saju,
I'am located in Tunesia ( North of Africa)
Thank you very much for your technical support the device is 250 km far a way from my office,
I try this week to travel in the clinic where are the BV 25 and do what you tell me after that I can give you more informations.

Best regards
Ben

Log inor Register

to rate and post a comment

saju bhaskaran

Re: BV25 PROBLEM WITH THE MEMORY

January 20, 2008 11:34

Dear Ben
under the posting where we were interacting found to be deleted from dotmed.????
Please let me know i posted a reply on further checks did u see my posting on how to go about the problem.please reply.are u base din US.we have plans to start our operations in US.
with regards

Log inor Register

to rate and post a comment

saju bhaskaran

Re: BV25 PROBLEM WITH THE MEMORY

January 21, 2008 08:02

Dear Ben

there will be one cable going to memory from bottom to i/p of the
memory.and two cables coming from the memory to the bottom
board.(if i am right)
1)swap two cables coming from the memory to bottom board and see
any change.
2)disconnect both the cables from memory to bottom board and see
whether live image is coming in the first monitor .
let me know the status.

Log inor Register

to rate and post a comment