C7 probe for Acuson - problem

(1)

Tzanko Mondeshki

C7 probe for Acuson - problem

August 16, 2006 08:43

I have an Acuson XP machine with ser.No 16107. The machine work perfectly. I buy this week C7 probe but when insert to the machine it not work and appear note Software for left transducer has not been licensed. Please remove left transducer. Do you know where the problem may be? I'd really appreciate your help Thank Tzanko Mondeshki e-mail - tz.m@abv.bg

Log inor Register

to rate and post a comment

(38)
(1)
(16)

Felix Donoghue

Re: C7 probe for Acuson - problem

August 17, 2006 12:01

The software for the operation of the C7 probe has not been installed, you would have to install the correct software to operate this probe.
felix

Log inor Register

to rate and post a comment

Jim Carr

Re: C7 probe for Acuson - problem

August 17, 2006 01:45

There are two things needed in order to operate a transducer on the Acuson XP and Aspen systems. The microcode for that transducer must be installed on the Memex board, and the software must be enabled by the proper encryption codes. On the 128XP, you can see the transducers that are enabled by following this procedure:
1. Depress and hold the Test, Code, Shift and V keys (Test is under the monitor behind the little door).
2. Use Trackball or keyboard to highlight 102 from the menu. Use Softkey to enter Select This will display the current Configuration.
3. Use Softkey to enter Page to get to list of transducers that are enabled. (There may be more than one Page.)

If the C7 is listed, then the microcode may be missing or it could be a connector problem on the transducer.

Jim Carr
Sonora Medical Systems

Log inor Register

to rate and post a comment

Michael H Badell

Re: C7 probe for Acuson - problem

August 17, 2006 05:50

Dear Mr. Mondeshki:
Each Acuson system has software licenses to run the probes. If your system is not licensed to run a particular probe you will get the message that you saw. Find out about paying for a license to run the probe.
Sincerely
Michael H. Badell
President
Accurate Medical Diagnostics
87A Island Drive #124
Alameda, CA 94502
(877)-548-4538 Toll Free
(510) 523-1432 Telephone
(800) 719-8875 Toll Free Fax
(510) 523-1620 Fax
www.AMDworldwide.com