FranciscoG

Can not do a STORELOG, issue with the octane computer

May 27, 2015 07:13

Hi guys,

I've got a LightSpeed Plus and everytime I want to do a STORELOG, it shows this message:

_____
PANIC: CPU 1: KERNEL FAULT
PC: 0xa80000002021f130 ep: 0xffffffffffffb898
EXC code: 128,'Software detected SEGV '
Bad addr: 0x0, cause: 0x8

Log inor Register

to rate and post a comment

(12)
(3)

Luis Santos

re: Can not do a STORELOG, issue with the octane computer

May 28, 2015 01:31

Reinstall software on new hard drives.

Check the SYSLOG and look for hard drive errors

1 person liked this
Log inor Register

to rate and post a comment

(140)
(35)

FranciscoG

re: re: Can not do a STORELOG, issue with the octane computer

March 09, 2018 03:52

Thanks Luis, I did it last time!

Log inor Register

to rate and post a comment

(32)
(4)

Tiorico_86

re: Can not do a STORELOG, issue with the octane computer

May 28, 2015 01:31

You must to run storelog with applications down from root...

1 person liked this
Log inor Register

to rate and post a comment

(140)
(35)

FranciscoG

re: re: Can not do a STORELOG, issue with the octane computer

March 09, 2018 03:53

Thanks Tiorico_86 I never could and can perform the storelog with applications down. Is there any other way to free disk space without needing to LFC software?

Log inor Register