Carestream 140 cr reader error message

JR Geis

Carestream 140 cr reader error message

November 28, 2013 02:24

CR reader would not accept the cassettes and gave the error message Error #134 Plate didn't reach W0 sensor. Reading the manual, the W0 Sensor is what drives the rollers to pull the screen into the machine. The available repair manual discusses where that sensor is and how to reach it, but does not explain what the error message is, or how to fix it. Appreciate you help

Log inor Register

to rate and post a comment

(42)
(7)

xtekvet

re: Carestream 140 cr reader error message

November 29, 2013 08:47

This is normally a transport issue. The plate is not entering far enough into the drum to get to the sensor. There are several things that can cause this. weak springs on the rollers or bad roller which do not cause enough friction to push the screen in. First thing I would do is clean the rollers and move the springs to a hole that gives them more tension. These should help for a while but long term you will have to replace rollers. If you need more info email me and I can send you which parts need attention and possible replacement.

1 person liked this
Log inor Register

to rate and post a comment

(2)

JR Geis

re: re: Carestream 140 cr reader error message

September 08, 2014 01:53

Thanks for the help. Apologies for late reply.

Log inor Register

to rate and post a comment

(1)
(1)

Hubert Baroso

re: Carestream 140 cr reader error message

December 01, 2013 07:51

There is 80% probability that your CR 140 is not broken, Please if the error happens with only 1 cassette, check the other cassettes to see if work, If it is working with the other cassettes the problem is that you have a cassette without plate and you have one cassette with two plates.

1 person liked this
Log inor Register

to rate and post a comment

(2)

JR Geis

re: re: Carestream 140 cr reader error message

September 08, 2014 01:53

Thanks for the reply. It does not work with multiple plates.