Ampronix, a Top Master Distributor for Sony Medical, provides Sales, Service & Exchanges for Sony Surgical Displays, Printers, & More. Rely on Us for Expert Support Tailored to Your Needs. Email info@ampronix.com or Call 949-273-8000 for Premier Pricing.
ANDRZEJ BARANSKI
Carestream Classic CR problem
May 11, 2020 10:40
Classic CR scanner not extracting the plate from chassis.
to rate and post a comment
juan pablo amado
re: Carestream Classic CR problem
May 15, 2020 11:42
do you have error code?
to rate and post a comment
Kamil Bayram
re: re: Carestream Classic CR problem
May 25, 2020 11:04
951-12021
to rate and post a comment
juan pablo amado
re: re: re: Carestream Classic CR problem
May 26, 2020 10:47
I can´t find 12021 error,
12012 is:
Description: The PLATE did not move within the correct time.
Cause: A malfunction occurred in one or more of the following components:
• COIL BOARD
• SLOW SCAN AY
Action: 1. Check for an obstruction.
2. Do the checkout for the SLOW SCAN CONTROLLER BOARD A6 and COIL
BOARD A7
to rate and post a comment
Kamil Bayram
re: re: re: re: Carestream Classic CR problem
May 27, 2020 11:12
Thank you
to rate and post a comment
CARLOS HUMBERTO RODRIGUEZ
re: re: re: re: Carestream Classic CR problem
January 21, 2025 12:51
I HAVE THE SAME CODE , 12021 , DO YOU HACE SOLUTION?
to rate and post a comment