April 18, 2025 10:27

The digitizer turns on without errors, picks up the plate, but when trying to digitize (when the linear motion starts), it crashes. The error occurs when trying to move the linear accelerator in any direction. When the device is already in an error state, you can move the linear accelerator in any direction using the engineer's menu. Sometimes with error number 7. I replaced the controller board, but this did not work. Please help solve the problem. Maybe it's the flex cables?
I recorded a video of this problem.
https://youtu.be/43fOsi8zTos

Log inor Register

to rate and post a comment

(17)
(1)

Root

re: Carestream Vita Error

April 21, 2025 02:27

Hello,

I watched your video and I think that the system isn't detecting the left-hand sensor. On your video at 59s, the right-hand sensor is lit up, so the left-hand one should be too.