edward bartra

Cath lab table cushion cleanable or replace?

June 18, 2008 03:16

we have a ge cath lab in our warehouse and x-ray tube oil leaked onto the table cusion and got into the padding inside. Is this cleanable or do I have to find a new one?

Log inor Register

to rate and post a comment

(244)
(19)

Staate Hayward

Re: Cath lab table cushion cleanable or replace?

June 19, 2008 09:38

I would suggest that you get a new pad. If the oil has gotten through the supposed water(fluid) tight covering then other fluids may be able to get in, ie. Blood. The oil shouldn't be visible in the image, but if it got in, it can get out and that is another sterile area contamination concern.

Log inor Register

to rate and post a comment