Over 150 Missouri Auctions End Tomorrow 06/05 - Bid Now

Ceretom Software

(1)

Mel Milosevic

Ceretom Software

April 18, 2017 10:36

Does someone have any contacts for the Ceretom? I have a used Ceretom (2008) model. The hard drive was wiped out before purchase. I am in need of the compatible software to communicate between the Ceretom and laptop.

Any help or leads would be greatly appreciated!

Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

re: Ceretom Software

April 19, 2017 08:18

Did you see this document in what I sent to you
Setting up a new laptop to connect with Scanner

Workstation Settings
1. Log into system as Admin / Administrator (password is “password”)
2. Select PACs configuration
3. In Workstation Settings, set Scanner IP Config to 192.SN#.1 (wireless or 3 for hardwire).100 (note: SN = Serial Number of Ceretom)
4. Click “Apply”

Network Settings
1. Click Microsoft key (flag)
2. Select Control Panel
3. Select Network Connections
4. Find either the hardwire System icon or the wireless icon (which ever applies) and right click
5. Choose Properties
6. Scroll to the bottom of the connection list and highlight Internet Protocol (TCP/IP)
7. Select properties
8. Ensure “Use the following IP address” is checked
9. Input the following IP address: 192.SN#.1 (wireless or 3 for hardwire).10
10. Input the following Subnet Mask: 255.255.255.0
11. Select OK (this completes the wired connection setup, continue for wireless setup)
12. Next, for wireless on System Properties window that is still open, select “Wireless Networks” tab
13. Select “View Wireless Connections”
14. Find the system SN# icon and double click it to connect
15. Next select “Change Advanced Settings” from related tasks list
16. Select the “Wireless Networks” tab once again
17. Under the Preferred Networks list, make sure that system SN# is only icon that appears (System SN# represents the actual scanner being used, remove all others).
18. Click OK and then close out all open windows to return to the run screen.
19. Log out as ADMIN and log in as operator.
20. View the status to ensure connection has occurred.
21. If asked for an encryption key, use the following: NeuroLogica# (where # is the last two numbers of the system serial number).

Computer Settings
1. Ensure that the Wireless” switch is turned on in the front of the laptop (if using a wireless connection to the scanner)

Log inor Register

to rate and post a comment

(332)
(37)
(60)

Gary Benitez

re: re: Ceretom Software

October 25, 2021 11:03

Hi Lenny,

Do you a copy of the service manual.
Also, I have a customer with an issue in Peru.
Do you have any contacts is Peru who can service it?
Please let me know.
Thanks,

Log inor Register

to rate and post a comment

aespy

re: re: Ceretom Software

January 24, 2023 11:46

Hi Lenny….could not figure your PW above for the Administrator. Can you resend. Thanks

Log inor Register