Checksum errors 7,9,10,11 during calibration

(1)

Muhammad Zubair

Checksum errors 7,9,10,11 during calibration

July 13, 2020 11:49

Hi Engineers
We were doing calibrations on GE AMX4 X-ray machine. It is not showing other calibration options instead of CAL DRIVE HANDLE. Other options in the menu are Checksum Errors 7,9,10,11.. Please let us know how we can recalibrate voltmeter and generator to get rid of these checksum errors.

Log inor Register

to rate and post a comment

(423)
(40)
(5)

Juan de la Cruz Martinez

re: Checksum errors 7,9,10,11 during calibration

July 14, 2020 10:59

Are you doing the complete calibration? All those errors are lack of calibration

Log inor Register

to rate and post a comment

(127)
(16)
(2)

bashir ahmad

re: Checksum errors 7,9,10,11 during calibration

July 14, 2020 11:00

Normally checksum error are caused by store data or cal files ist check the battery on cpu board and than complete re-calibrate the system.

Log inor Register

to rate and post a comment

(28)
(1)
(1)

Howard Karp

re: Checksum errors 7,9,10,11 during calibration

July 20, 2020 11:16

Please confirm if the System is Failing Initial Turn On TEST 4

Log inor Register

to rate and post a comment

(28)
(1)
(1)

Howard Karp

re: Checksum errors 7,9,10,11 during calibration

July 20, 2020 11:16

Are the System Batteries in Good Condition? When were they changed?

Log inor Register

to rate and post a comment

(28)
(1)
(1)

Howard Karp

re: Checksum errors 7,9,10,11 during calibration

July 20, 2020 11:16

Please confirm if the System is Failing Startup Test # 4 ? What condition are the Batteries? When were they changed last?

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Your one-stop source for all Pre-Owned Medical Imaging Equipment needs. 📞 305-234-0084

We supply high-quality refurbished and used imaging equipment and parts to hospitals, imaging centers, and brokers worldwide. We pay top dollar and offer installation, deinstallation, service, and transport. Se habla español info@medilabglobal.com

DOTmed Certified DOTmed 100 stats

Upcoming Auction Events

Cbrx2sales
TIMED Ending Soon
Jun 05, 2026 (157 lots)