This issue is related to dirty optics, misaligned optics, or burnt out optics - usually the KTP crystal itself, which requires an experienced tech to align properly. You will also need an Iridex or Cutera service card, depending on the software version of your device. Your lamps could also be out.
Ophthalmic-focused expertise in YAG systems including repair, maintenance and calibration handled with precision by trained professionals. Request service • sales@laserlocators.com • 1-877-924-2020
Code 46 Lasercope Gemini
March 22, 2021 10:31
I have an Gemini dualwave 1064nm,532 KTP, when the equipment starts, on the screen appears code 46.
does Someone know what can i do to solve it?
to rate and post a comment
Mickey A. Couvertier
re: Code 46 Lasercope Gemini
March 23, 2021 11:10
This issue is related to dirty optics, misaligned optics, or burnt out optics - usually the KTP crystal itself, which requires an experienced tech to align properly. You will also need an Iridex or Cutera service card, depending on the software version of your device. Your lamps could also be out.