I have a GE Sigma Exite II MRI that aleatory shows Corduroy artifacts. The strange with this is that a problem to move the cough appears just in this artifacts occurrence.
Due this fact I don´t want to proceed for corduroy troubleshooting yet assuming there is something I am not seen yet.
Jose Morillo
CODUROY ARTIFACT + COUGH MOVEMENT PROBLEMS
March 11, 2018 10:20
Dear colleagues:
I have a GE Sigma Exite II MRI that aleatory shows Corduroy artifacts. The strange with this is that a problem to move the cough appears just in this artifacts occurrence.
Due this fact I don´t want to proceed for corduroy troubleshooting yet assuming there is something I am not seen yet.
Any Clue?
Best regards
Jose
to rate and post a comment
Jim Foros
re: CODUROY ARTIFACT + COUGH MOVEMENT PROBLEMS
April 17, 2018 09:43
Check horizontal drive motor brushes at back of rear pedestal
to rate and post a comment
Jose Morillo
re: re: CODUROY ARTIFACT + COUGH MOVEMENT PROBLEMS
April 18, 2018 09:58
Thanks Jim. We already found the problem in decoder box. Problem is solved.
Best regards
to rate and post a comment