There is a possibility that this error is caused by the CPU board. Try pressing down all IC's which are on sockets.
Most UP5000 systems start suffering from ghost errors due to unstable signals on the cpu.
Ophthalmic-focused expertise in YAG systems including repair, maintenance and calibration handled with precision by trained professionals. Request service • sales@laserlocators.com • 1-877-924-2020
NordicInnoTech
Coherent Ultrapulse 5000 error
October 02, 2013 11:07
Any information on error code 23 during start-up.
to rate and post a comment
Lachman Aswani
re: Coherent Ultrapulse 5000 error
October 02, 2013 07:01
FAULT 23 - NO FIRE - (Cleared by pressing ready, active in service mode). The NO FIRE signal has been asserted.
==>> Bad tube
to rate and post a comment
Wouter Kelderman
re: Coherent Ultrapulse 5000 error
October 03, 2013 08:25
There is a possibility that this error is caused by the CPU board. Try pressing down all IC's which are on sockets.
Most UP5000 systems start suffering from ghost errors due to unstable signals on the cpu.
to rate and post a comment
Tadamasa Teratani
re: re: Coherent Ultrapulse 5000 error
December 02, 2016 09:15
"Please call your service representatives as soon [11]" appeared at boot up.
what mean the error code 11?
The ultrapulse dose not recognize deep and active FX scanners.
If I go into service mode, what should I do for fix them?
Please help !
kagiya@gmail.com
to rate and post a comment
Scott Scholl
re: re: re: Coherent Ultrapulse 5000 error
August 06, 2018 08:15
would anyone know what error code 52 is on a 5000c pulse
to rate and post a comment
Douglas Green
re: re: re: re: Coherent Ultrapulse 5000 error
October 30, 2018 02:53
52= no water flow....check water level, and sometimes flow switch sticks