Thomas,
That error code is actually a 5.0. The first step is to check the A4 board for a blown fuse. It means the power was measured too low during the self check at power-up. We also have the service manual if you havent already gotten it.
Thanks
Rich
SE Medical Systems.
Thomas Marr
Conmed Excalliber plus E.S.U. error code
January 14, 2013 03:56
Hello,
We have a excaliber plus with an error code 50
No manual . Need some help,
Thanks in advance
to rate and post a comment
Alexandre Santos
re: Conmed Excalliber plus E.S.U. error code
January 14, 2013 06:21
Hello,
I have the service manual. Let me know your email address.
Best Regards,
Alexandre Santos.:
to rate and post a comment
Thomas Marr
re: Conmed Excalliber plus E.S.U. error code
January 14, 2013 06:30
Thanks Alexander,
Ive got a ton of manuals here.Ifyouneed something let me know, Thanks Thomas arrowheadbio@charter.net
to rate and post a comment
Richard Johnson
re: Conmed Excalliber plus E.S.U. error code
January 14, 2013 06:31
Thomas,
That error code is actually a 5.0. The first step is to check the A4 board for a blown fuse. It means the power was measured too low during the self check at power-up. We also have the service manual if you havent already gotten it.
Thanks
Rich
SE Medical Systems.
to rate and post a comment
Thomas Marr
re: re: Conmed Excalliber plus E.S.U. error code
January 14, 2013 09:28
Hi Rich I'll give it a shot thanks
Tom
to rate and post a comment