continental tm30 multimple flags

ROBERTO DIAZ BEDOLLA

continental tm30 multimple flags

November 29, 2011 08:08

HI. I HAVE A CONTINENTAL TM30 XRAY GENERATOR, CONTROL IS DISPLAYING AT LEAST 5 FLAGS LIKE 64 95 ETC,... I HAVE THE SERVICE MANUAL IT SEEMS A COMUNICATION ERROR BUT DESPITE OF THIS I DON'T KNOW IF IT'S NECESARY TO CHANGE ANY CARD AND WHICH ONE?... ALL ANSWERS MAY HELP ME THX...

Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

Re: continental tm30 multimple flags

November 29, 2011 05:41

Check your +5VDC power supplies, if one is low below 4.85 it will prevent the circuit from powering up and it will interrupt the CAN bus flagging several errors depending on what circuit is not coming up

1 person liked this
Log inor Register

to rate and post a comment

ROBERTO DIAZ BEDOLLA

Re: continental tm30 multimple flags

November 30, 2011 10:20

thx I will check them

Log inor Register

to rate and post a comment

salah Gamal

re: Re: continental tm30 multimple flags

March 27, 2014 01:06

HI.we have a problem for continental TM30 9 Flags filament overload

Log inor Register

to rate and post a comment

salah Gamal

re: Re: continental tm30 multimple flags

March 27, 2014 01:06

we have a problem for continental TM30 9 Flags filament overload

Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

re: re: Re: continental tm30 multimple flags

March 28, 2014 05:37

9 21 1,2 FILAMENT OVERLOAD ERROR
Exposure terminated by 5 consecutive (5ms) filament overload conditions,
actual filament current exceeds filament max (D/A IC10 pin 2). Filament max
set to 110% for filament feedback control and 120% for mA feedback control.
Most likely cause: Incorrect calibration, preheat offset numbers are too
low (can also be too high, this can happen on very low mA settings) See “Filament
Overload & mA Overload Error Detection Circuit” on page 80

I can send you a manual that might help

Log inor Register

to rate and post a comment

(3)

Willy Ramos

re: Re: continental tm30 multimple flags

March 28, 2016 05:31

Hi;;; You have a operators manual and installation,,, of continental TM30... could you please send me or tell me where I can find ... greetings Nelson Ramos

Log inor Register

to rate and post a comment

(42)
(3)

Doug Lally

Re: continental tm30 multimple flags

November 29, 2011 10:19

You want to talk to a man named Hugh Hudson. He retired from Hologic and gives support from His home.He is great and only charges you, if He gets in deep with you and your problem. His number is 971-404-4634 and is in Portland Oregon.I usually give him some beer money,having him as a contact is priceless.He can also help with Bennett and other older units. tell him Doug Lally from Medical Imaging said hello!

1 person liked this
Log inor Register

to rate and post a comment

ROBERTO DIAZ BEDOLLA

Re: continental tm30 multimple flags

November 30, 2011 10:20

thanks I'll probably get in touch with him!

Log inor Register

to rate and post a comment

(84)
(11)
(6)

Galen Hiveley

Re: continental tm30 multimple flags

December 10, 2011 09:39

That's good info. I knew Hugh did support on the Bennetts, but I didn't know he also did support on TM generators.

Log inor Register

to rate and post a comment

(3)

Willy Ramos

re: Re: continental tm30 multimple flags

March 28, 2016 05:31

Hi;;; You have a operators manual and installation,,, of continental TM30... could you please send me or tell me where I can find ... greetings Nelson Ramos

Log inor Register

to rate and post a comment

(3)

Willy Ramos

re: continental tm30 multimple flags

March 28, 2016 05:31

Hi;;; You have a operators manual and installation,,, of continental TM30... could you please send me or tell me where I can find ... greetings Nelson Ramos

Log inor Register

to rate and post a comment

(23)
(6)
(1)

Raymundo Moreno

re: continental tm30 multimple flags

May 31, 2016 10:35

Mr. Lenny, I have several errors: First, Invalid data front panelsystem F199 #94 factors will exceed tube heat linit. Only Small focus works. Then (now) both small and large does not work. IT seems to be the NVRAM is gone. Do you have support to have a copy? I also need the service manual. Please help me. If yes, tell me your data.

Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

re: re: continental tm30 multimple flags

May 31, 2016 12:39

yes it looks like the battery failed i can send you the service manuals and procedure on how to configure. it would be a lot easier if you had a back up but if not you can always manually configure the system takes longer.