HI. I HAVE A CONTINENTAL TM30 XRAY GENERATOR, CONTROL IS DISPLAYING AT LEAST 5 FLAGS LIKE 64 95 ETC,... I HAVE THE SERVICE MANUAL IT SEEMS A COMUNICATION ERROR BUT DESPITE OF THIS I DON'T KNOW IF IT'S NECESARY TO CHANGE ANY CARD AND WHICH ONE?... ALL ANSWERS MAY HELP ME THX...
Check your +5VDC power supplies, if one is low below 4.85 it will prevent the circuit from powering up and it will interrupt the CAN bus flagging several errors depending on what circuit is not coming up
9 21 1,2 FILAMENT OVERLOAD ERROR
Exposure terminated by 5 consecutive (5ms) filament overload conditions,
actual filament current exceeds filament max (D/A IC10 pin 2). Filament max
set to 110% for filament feedback control and 120% for mA feedback control.
Most likely cause: Incorrect calibration, preheat offset numbers are too
low (can also be too high, this can happen on very low mA settings) See “Filament
Overload & mA Overload Error Detection Circuit” on page 80
Hi;;; You have a operators manual and installation,,, of continental TM30... could you please send me or tell me where I can find ... greetings Nelson Ramos
You want to talk to a man named Hugh Hudson. He retired from Hologic and gives support from His home.He is great and only charges you, if He gets in deep with you and your problem. His number is 971-404-4634 and is in Portland Oregon.I usually give him some beer money,having him as a contact is priceless.He can also help with Bennett and other older units. tell him Doug Lally from Medical Imaging said hello!
Hi;;; You have a operators manual and installation,,, of continental TM30... could you please send me or tell me where I can find ... greetings Nelson Ramos
Hi;;; You have a operators manual and installation,,, of continental TM30... could you please send me or tell me where I can find ... greetings Nelson Ramos
Mr. Lenny, I have several errors: First, Invalid data front panelsystem F199 #94 factors will exceed tube heat linit. Only Small focus works. Then (now) both small and large does not work. IT seems to be the NVRAM is gone. Do you have support to have a copy? I also need the service manual. Please help me. If yes, tell me your data.
yes it looks like the battery failed i can send you the service manuals and procedure on how to configure. it would be a lot easier if you had a back up but if not you can always manually configure the system takes longer.
BMI-PMI services nearly every make, model, and modality through our nationwide team of nearly 200 local service engineers. Flexible contracts, 24/7 remote monitoring, and ISO 13485 certification. www.brownsmedicalimaging.com / 800-701-9729
ROBERTO DIAZ BEDOLLA
continental tm30 multimple flags
November 29, 2011 08:08
HI. I HAVE A CONTINENTAL TM30 XRAY GENERATOR, CONTROL IS DISPLAYING AT LEAST 5 FLAGS LIKE 64 95 ETC,... I HAVE THE SERVICE MANUAL IT SEEMS A COMUNICATION ERROR BUT DESPITE OF THIS I DON'T KNOW IF IT'S NECESARY TO CHANGE ANY CARD AND WHICH ONE?... ALL ANSWERS MAY HELP ME THX...
to rate and post a comment
Lenny Place
Re: continental tm30 multimple flags
November 29, 2011 05:41
Check your +5VDC power supplies, if one is low below 4.85 it will prevent the circuit from powering up and it will interrupt the CAN bus flagging several errors depending on what circuit is not coming up
to rate and post a comment
ROBERTO DIAZ BEDOLLA
Re: continental tm30 multimple flags
November 30, 2011 10:20
thx I will check them
to rate and post a comment
salah Gamal
re: Re: continental tm30 multimple flags
March 27, 2014 01:06
HI.we have a problem for continental TM30 9 Flags filament overload
to rate and post a comment
salah Gamal
re: Re: continental tm30 multimple flags
March 27, 2014 01:06
we have a problem for continental TM30 9 Flags filament overload
to rate and post a comment
Lenny Place
re: re: Re: continental tm30 multimple flags
March 28, 2014 05:37
9 21 1,2 FILAMENT OVERLOAD ERROR
Exposure terminated by 5 consecutive (5ms) filament overload conditions,
actual filament current exceeds filament max (D/A IC10 pin 2). Filament max
set to 110% for filament feedback control and 120% for mA feedback control.
Most likely cause: Incorrect calibration, preheat offset numbers are too
low (can also be too high, this can happen on very low mA settings) See “Filament
Overload & mA Overload Error Detection Circuit” on page 80
I can send you a manual that might help
to rate and post a comment
Willy Ramos
re: Re: continental tm30 multimple flags
March 28, 2016 05:31
Hi;;; You have a operators manual and installation,,, of continental TM30... could you please send me or tell me where I can find ... greetings Nelson Ramos
to rate and post a comment
Doug Lally
Re: continental tm30 multimple flags
November 29, 2011 10:19
You want to talk to a man named Hugh Hudson. He retired from Hologic and gives support from His home.He is great and only charges you, if He gets in deep with you and your problem. His number is 971-404-4634 and is in Portland Oregon.I usually give him some beer money,having him as a contact is priceless.He can also help with Bennett and other older units. tell him Doug Lally from Medical Imaging said hello!
to rate and post a comment
ROBERTO DIAZ BEDOLLA
Re: continental tm30 multimple flags
November 30, 2011 10:20
thanks I'll probably get in touch with him!
to rate and post a comment
Galen Hiveley
Re: continental tm30 multimple flags
December 10, 2011 09:39
That's good info. I knew Hugh did support on the Bennetts, but I didn't know he also did support on TM generators.
to rate and post a comment
Willy Ramos
re: Re: continental tm30 multimple flags
March 28, 2016 05:31
Hi;;; You have a operators manual and installation,,, of continental TM30... could you please send me or tell me where I can find ... greetings Nelson Ramos
to rate and post a comment
Willy Ramos
re: continental tm30 multimple flags
March 28, 2016 05:31
Hi;;; You have a operators manual and installation,,, of continental TM30... could you please send me or tell me where I can find ... greetings Nelson Ramos
to rate and post a comment
Raymundo Moreno
re: continental tm30 multimple flags
May 31, 2016 10:35
Mr. Lenny, I have several errors: First, Invalid data front panelsystem F199 #94 factors will exceed tube heat linit. Only Small focus works. Then (now) both small and large does not work. IT seems to be the NVRAM is gone. Do you have support to have a copy? I also need the service manual. Please help me. If yes, tell me your data.
to rate and post a comment
Lenny Place
re: re: continental tm30 multimple flags
May 31, 2016 12:39
yes it looks like the battery failed i can send you the service manuals and procedure on how to configure. it would be a lot easier if you had a back up but if not you can always manually configure the system takes longer.