I am installing a CONTINENTAL TM30 Xray device but have a problem, in the operator console ther is a permanent HOLD signal which inhibits any xray generation atempts, we can select a techinique but when press the preparation switch a 84 flag apears in the LCD display. The description of the falg 81 is that the tube would be to hot and have to allow cool but we didnt made even one shoot so the xray tube is copletely cold. We already connected all the cables and interlocks. Does any one knows wh the HOLD indicator is always present? could some one have more information about?
Thansk beforehand
Flag 84 = FACTORS WILL EXCEED TUBE HEAT LIMIT
Flag cleared by reducing tube power or by allowing tube to cool.
* Is it possible that there is an open tube thermal interlock?
email me if you want some training documentation.
--------------------------
Wayne M. Horsman, VP
Columbia Imaging, Inc.
9050 Red Branch Rd., Ste. O
Columbia, MD, USA 21045
www.columbiaimaging.com
wayne@columbiaimaging.com
Definitely sounds like tube thermal interlock. Also look at tube programming. If you have incorrect tube programmed and you set techniques to where it's above tube rating you will get the error. Check your ABC's
BMI-PMI services nearly every make, model, and modality through our nationwide team of nearly 200 local service engineers. Flexible contracts, 24/7 remote monitoring, and ISO 13485 certification. www.brownsmedicalimaging.com / 800-701-9729
Ruben Cadima
CONTINENTAL TM30 PROBLEM
June 08, 2014 08:48
Hello dears
I am installing a CONTINENTAL TM30 Xray device but have a problem, in the operator console ther is a permanent HOLD signal which inhibits any xray generation atempts, we can select a techinique but when press the preparation switch a 84 flag apears in the LCD display. The description of the falg 81 is that the tube would be to hot and have to allow cool but we didnt made even one shoot so the xray tube is copletely cold. We already connected all the cables and interlocks. Does any one knows wh the HOLD indicator is always present? could some one have more information about?
Thansk beforehand
to rate and post a comment
Wayne Horsman
re: CONTINENTAL TM30 PROBLEM
June 09, 2014 09:12
Flag 84 = FACTORS WILL EXCEED TUBE HEAT LIMIT
Flag cleared by reducing tube power or by allowing tube to cool.
* Is it possible that there is an open tube thermal interlock?
email me if you want some training documentation.
--------------------------
Wayne M. Horsman, VP
Columbia Imaging, Inc.
9050 Red Branch Rd., Ste. O
Columbia, MD, USA 21045
www.columbiaimaging.com
wayne@columbiaimaging.com
to rate and post a comment
Galen Hiveley
re: re: CONTINENTAL TM30 PROBLEM
December 20, 2017 03:37
Hello Wayne,
I'm trying to find a console for a TM30.
Thanks,
Galen
Email: galen@xraypartsdepot.com
to rate and post a comment
Juan Antonio Ramirez Ramirez
re: CONTINENTAL TM30 PROBLEM
December 21, 2017 09:18
Do you have another keyboard your problem must be keyboard
to rate and post a comment
Galen Hiveley
re: re: CONTINENTAL TM30 PROBLEM
December 21, 2017 12:10
Do not have another keyboard. Trying to find the entire console.
to rate and post a comment
Ronald Swapp
re: CONTINENTAL TM30 PROBLEM
January 12, 2018 09:36
Definitely sounds like tube thermal interlock. Also look at tube programming. If you have incorrect tube programmed and you set techniques to where it's above tube rating you will get the error. Check your ABC's