CONTINENTAL TM-40 HELP NEEDED

(77)
(1)
(4)

eduardo medina

CONTINENTAL TM-40 HELP NEEDED

March 18, 2010 09:35

Someone who have expirence on that that Generator, is a used system on installation start to make radiation but inmediatly cut sending an error flag# 5.After shecking start to send flag # 53 surge contactor.
Any idea?
Eduardo

Log inor Register

to rate and post a comment

(26)
(4)

Danial Rao

Re: CONTINENTAL TM-40 HELP NEEDED

March 18, 2010 04:57

there a red color push button on the back side of the cabinet, that is for on/off the inverter located just up side the 3 phase wire conection, i think it will be off.

Log inor Register

to rate and post a comment

(77)
(1)
(4)

eduardo medina

Re: CONTINENTAL TM-40 HELP NEEDED

March 20, 2010 11:46

Thank you for answer
In fact is ON I observed on top of cabinet a DS1 turn ON aparently is feedback interlock. The Main contactor(M3) and M1(Cr) not enabled. Aparently the 24Vac return is not present. But I dont know where start that signal if born on KVC board or Field I/0 Board.
Do you know?Any idea? What do you consider to change?
Eduardo

Log inor Register

to rate and post a comment

Doug Gray

Re: CONTINENTAL TM-40 HELP NEEDED

March 26, 2010 11:47

That 24V comes form the I/O board in the card cage
There is a isolated voltage regulator on that board. It has HDF-2405S marked on it. It is probably bad.

doug@allstarxray.com

Log inor Register

to rate and post a comment