Continental TM-40

October 14, 2011 09:19

When i make exposure the machine show error flag #14 KV unbalanced, i already check the HT cables and X-ray tube I need some tip in order to help me .I will appreciate any advice. Thank you Ramon Diaz

Log inor Register

to rate and post a comment

(2)
(1)

Wesley Moore

Re: Continental TM-40

October 15, 2011 09:05

Hi Ramon
Check the inverter circuit in the generator. Specifically the driver transistors.
And I seem to remember that that error can be caused by a bad driver transistor in the filament circuit too.

good luck
Carlton

Log inor Register

to rate and post a comment

(21)
(2)

Scott Deason

Re: Continental TM-40

October 19, 2011 02:59

YOU NEED TO CHECK KV WITH A VOLTAGE DIVIDER.VERIFY THAT THE POS. AND NEG. ARE EQUAL,I HAVE HAD BAD HIGH VOLTAGE TRANSFORMER...SCOTT

Log inor Register

to rate and post a comment

(7)
(1)
(2)

Ramon Diaz

Re: Continental TM-40

October 20, 2011 12:22

Hi Scott. Do you think the HT transformer could be bad , did you have the same error flag before, what was the cause? thank you for your help

Log inor Register

to rate and post a comment

(84)
(11)
(6)

Galen Hiveley

Re: Continental TM-40

October 23, 2011 11:50

The hi-voltage transformer on these generators had a high rate of failure. Replacements are difficult to find.

Log inor Register

to rate and post a comment

(21)
(2)

Scott Deason

Re: Continental TM-40

November 01, 2011 06:07

RAMON I HAVE HAD THIS BEFORE, YOU CAN MEASURE THE ACTUAL KV WITH A DIVIDER TANK .LOOK AT PLUS AND NEG> WAVE FORM THEY MUST BE EQUAL IF NOT THE TANK WOULD NEED TO BE REPLACED..I HAVE ATTEMPTED TO ISOLATE PROBLEM IN THE TRANSFORMER AND BELIEVE IT TO BE THE WINDINGS,BUT NOT SURE.
GOOD LUCK,SCOTT

Log inor Register

to rate and post a comment

(42)
(3)

Doug Lally

Re: Continental TM-40

November 29, 2011 10:19

Hugh Hudson,His number is 971-404-4634. He is retired tech support from Hologic He's the best you may need to give him a P.O. but usually he can tell you whats wrong or what to do in 5 min. If it gets really involved then you may need to get him a P.O.