Over 150 Missouri Auctions End Tomorrow 06/05 - Bid Now

Contour Boot Problem

Tags: MRI, GE

Posted by an Anonymous User

Contour Boot Problem

February 25, 2013 01:34

I got troubles with a Starup GE MRI Contour 3, (0.5T),
problem started loosen DATE/TIME
now it does not star up at all
it lost Ethernet address and host ID

"Sun Ultra 2 UPA/SBus(2X UltraSPARC 168 MHz), Keyboard present
OpenBoot 3.19, 256 MB memory installed, serial 16777215
Ethernet address ff:ff:ff:ff:ff:ff, Host ID: ffffffff

The IDPROM contents are invalid

Firmware password:
Type boot, go (continue), or login (command mode)
>"

Any solution about it?
thank you

Log inor Register

to rate and post a comment

(8)
(2)

Ing. Alejandro Alhadeff

re: Contour Boot Problem

February 25, 2013 07:59

Your NVRAM in the CPU inside the console is already bad. You have to replace it and restore the original HostId on the new NVRAM. I wish you already did a backup of the original HostId, otherwise you are in a little big problem. You had to replace the NVRAM when the date and time started to show you problems. Replace the NVRAM and restore the HostId and the MR will boot again. Regards

Log inor Register

to rate and post a comment

(70)
(12)

Julio c. Gomez

re: Contour Boot Problem

February 26, 2013 08:21

Hello Anonymous.

sun ultra 2
model: (sun4u)
NVRAM location on motherboard: U0205
NVRAM type: M48T59Y

first change the old NVRAM for one new

use the mkp command at the (ok) pront for change or restore you original HostID.

for make easy above please insert a Solaris 5.1 CD software and type: boot - cd from (ok)

mount the solaris kernel and see the ethernet log file for extract the mac address you will use this for restore you system.

good luck.
bye.

Julio C. Gomez
pruebadecampo@yahoo.es



Log inor Register

to rate and post a comment

stats
DOTmed text ad

Philips Achieva 1.5T MRI Spare Parts Available Looking for Philips Achieva 1.5T MRI Parts?

We offer reliable & cost-effective spare parts for the Philips Achieva 1.5T MRI system helping you minimize downtime & keep operations smooth. Wide inventory available. Quick turnaround time. Expert support. Worldwide delivery. Contact Limor@elsmed.com

DOTmed Certified stats

Upcoming Auction Events

Cbrx2sales
TIMED Ending Soon
Jun 05, 2026 (157 lots)