coolpak 6000 compressor oil cct fail error

Tags: MRI, COPELAND

Posted by an Anonymous User

coolpak 6000 compressor oil cct fail error

December 26, 2015 10:10

I get an error from Coolpak 6000 compressor.I checked all the parameters, there is no problem on chiller,no problem on oil level.Howeever,it works for a while and then oil temperature gets higher and I get an "oil cct fail" error.If anyone has an opinion about this error, please be so kind to share your views & commented.Thanks in advance.

Log inor Register

to rate and post a comment

(13)

(12)

Jody Rotenizer

re: coolpak 6000 compressor oil cct fail error

December 27, 2015 10:24

The heat exchanger inside of the may be clogged and a good way to try to clear it is to reverse the incoming water lines. With that said it is usually the incoming water from either temp or flow issues.

Log inor Register

to rate and post a comment

(22)
(4)

Fedor Mikolakhin

re: coolpak 6000 compressor oil cct fail error

December 27, 2015 10:26

Basically, this error indicates that oil cooling circuit works not good enough. It may be due to compressor capsule itself overheating, due to oil leak, due to congested heat exchanger, etc.
If everything is fine with cooling (including heat exchanger cleaned from dirt) and compressor was never maintained within 8-10 years, it's standard error for aged Coolpak6000 and HC-10 compressors with vertical compressor capsule.
You can try to add some oil and it might help for a while (if your adsorber is new and displacer have to be replaced soon as it may induce much faster oil contamination of gas lines and coldhead), but usually such compressors need full maintenance, including cleaning and full oil exchange.
Every cryoservice guy would recommend that.

Right on site I'd
clean water heat exchanger first,
look for any possible oil leaks,
check resistance of compressor motor windings or checked compressor capsule somehow
and then
if compressor is aged and oil in sight glass is not sufficiently clean (i.e. colored brown, green, etc.) look for cryo-repair company.

1 person liked this
Log inor Register

to rate and post a comment

FIRAT KORKMAZ

re: re: coolpak 6000 compressor oil cct fail error

January 04, 2016 10:10

Hi Fedor, thanks for your comment & support.It helped much.

Log inor Register

to rate and post a comment

(110)
(10)
(5)

Wojciech Selwesiuk

re: coolpak 6000 compressor oil cct fail error

December 27, 2015 10:26

I think the coolant doesn't flow throught the compressor. Possible are: (1) pipes bended or internally blocked (2) heat exchanger is blocked. Remove water outlet and check real flow and temperature. Check also outlet water pipe. (I had blocked this pipe one time) In second case unit has to be replaced

Log inor Register

to rate and post a comment

(20)

Fabio Moreira

re: coolpak 6000 compressor oil cct fail error

November 21, 2018 10:01

Does any one has a service manual for the Collpak 6000 ?
If so, please send it to: fabio@novacorp-sa.com
Thanks

Log inor Register

to rate and post a comment

stats
DOTmed text ad

GE, Siemens, Philips, & Toshiba/Canon imaging parts, training & support!

DirectMed Imaging specializes in Equipment Sales, Parts, and Support. Parts are in stock and available for immediate delivery. In-room MRI Upgrades are available now. (855) 463-3727 or sales@directmedimaging.com

DOTmed Certified stats

Upcoming Auction Events

Cbrx2sales
TIMED Ending Soon
Jun 05, 2026 (157 lots)