CR 35 AGFA
September 24, 2015 07:14
I have a cr 35 have menssage lamp 3 off, check clean workink and appear again, somebody could help? lok for this service manual too please
to rate and post a comment
Mario E Sarduy La Paz
re: CR 35 AGFA
September 25, 2015 07:57
Hi Yojin Ramones. Do you speak spanish? because my native language is spanish and is more easy for me give an explanation. Anyway you can solve it very easy. Shut down your CR35 and open the right panel, you will see a panel with 2 connetors, disconnects both and the four screws and dismount this panel. You will see the lamps, open the closures and exchange the third lamp counting the left to the right for a 12 V 100 W LAMP. close the closures, put again the pannel and connet both connectors, fix the four screws and mount again the rigth panel. power on your system and it is all that you need to do. Regards
to rate and post a comment
Yojin Ramones
re: re: CR 35 AGFA
September 25, 2015 03:11
Mario thanks a lot, I just do it, and I could see all lamp on, and after that star the secuence an appear lamp off 3 defective error 25102.
to rate and post a comment
Lenny Place
re: CR 35 AGFA
September 25, 2015 11:14
I can send you the manual would need an email address
to rate and post a comment
Daniyar Batalov
re: re: CR 35 AGFA
August 22, 2017 07:23
send it manual for me, please. my mail is den9319@mail.ru
to rate and post a comment
Daniyar Batalov
re: re: CR 35 AGFA
August 22, 2017 07:24
Hello, can you send this manual for me, please? my mail is den9319@mail.ru
to rate and post a comment