Hi there Who know how i install external warning light pin connector
to rate and post a comment
Kiss Gábor
re: CRANEX TOME (SOREDEX)
October 24, 2018 08:41
Hi! As I see it get out a 5 V signal. Connector J804 (this is on the board N800) and external lights connector DIN3 Pin Signal Description 1 ISOL_EXP Active when exposure button pressed 2 PGND - 3 ISOL_RDY Active when unit is ready for exposures If you can mesure a GND from that you can mesoure a signal for exposure switch also. I hope that above information can help you. Best regards Gábor Kiss
Therapeutic Technologies closeout auctions take place on DOTmed.com. Bid today on 538 lots of medical equipment. Inspections may be done by appointment. Hours are Monday through Friday 9am-4pm Contact chrisnuytens1@gmail.com to schedule.
FANOS SPYROU
CRANEX TOME (SOREDEX)
October 15, 2018 08:51
Hi there
Who know how i install external warning light
pin connector
to rate and post a comment
Kiss Gábor
re: CRANEX TOME (SOREDEX)
October 24, 2018 08:41
Hi!
As I see it get out a 5 V signal.
Connector J804 (this is on the board N800)
and external lights connector DIN3
Pin Signal Description
1 ISOL_EXP Active when exposure button pressed
2 PGND -
3 ISOL_RDY Active when unit is ready for exposures
If you can mesure a GND from that you can mesoure a signal for exposure switch also.
I hope that above information can help you.
Best regards
Gábor Kiss