CT Activion 16 scan problem

June 14, 2012 04:37

Dear friends,

This is an Activion 16.

After formating all HDDs in RDD and Image disk initialization
any attempt to scan (Air Calibr, NRA, patient scan) leads
to the following error message:

An abnormal scan conditions was found.
Please restart the exam plan.

code [0xc0900011] dev[0x00140000] Pno [1]

Warm up could be done with no problem.
When scan starts the gatry makes one or more
rotations, starts x-ray explosion and stops
with the above message.

Any suggestion is appreciated

Alex
reutaav@inbox.ru

Log inor Register

to rate and post a comment

(9)
(3)

Gee Gicale

re: CT Activion 16 scan problem

June 14, 2012 05:59

Hi, Did you replaced 1 rdd disk or more? Have you tried Reconfigure RDD with parity.

Log inor Register

to rate and post a comment

(26)
(3)

Alexandr Reuta

re: re: CT Activion 16 scan problem

June 15, 2012 09:44

Thank you for your answer

No disk has been replaced. RDDUtil says that all disks are good. When we try to Reconfigure RDD with parity the system says that the RTM is not ready and waits too much time...

Log inor Register

to rate and post a comment

(9)
(3)

Gee Gicale

re: re: re: CT Activion 16 scan problem

June 15, 2012 03:25

Reconfigure RDD takes a lot of time, it ranges from 45 to 85 minutes. Try doing that again, hopefully you'll get a message " Rdd done"

Log inor Register

to rate and post a comment

(26)
(3)

Alexandr Reuta

re: re: re: re: CT Activion 16 scan problem

June 17, 2012 09:44

Thank you very much! We will try...

Log inor Register

to rate and post a comment

(1)

Jun Fang

re: CT Activion 16 scan problem

June 15, 2012 09:42

Some of the MHR pwb maybe broken, use the MHR diagnostic program to identify which is the bad one

Log inor Register

to rate and post a comment

(26)
(3)

Alexandr Reuta

re: re: CT Activion 16 scan problem

June 17, 2012 09:44

Thank you very much for your support! We will try this.

Log inor Register

to rate and post a comment

(88)
(36)

Anton Spaci

re: CT Activion 16 scan problem

June 30, 2012 11:43

access of RDD is not good
MHR FC could be faulty

1 person liked this
Log inor Register

to rate and post a comment

(26)
(3)

Alexandr Reuta

re: re: CT Activion 16 scan problem

July 01, 2012 09:07

thank you very much for your reply!
The problem was ubnormal currents from the generator.
RDD & PCBs are Ok

Log inor Register

to rate and post a comment

stats
DOTmed text ad

MRI & CT Support Worldwide. Buy Parts Fast with LASO.

Comprehensive medical imaging support. We supply essential parts & maintenance for most OEMs, buy/sell quality used equipment, and offer mobile rentals globally. Contact us: info@lasoimaging.com

stats

Upcoming Auction Events

Cbrx2sales
TIMED Ending Soon
Jun 05, 2026 (157 lots)