Ct Brain image getting Cut

Tags: CT, GE

(8)

Psmatr

Ct Brain image getting Cut

March 14, 2017 10:24

Hi There ,
The machine in question is a Hispeed Dual CTe , Images are getting cut while planning and doing a CT Brain .The rest of the scans i.e. KUP or chest etc are no issue. Any Idea what this might be?
Thanks in advance

Log inor Register

to rate and post a comment

(88)
(4)

Edurado Madeiros

re: Ct Brain image getting Cut

March 15, 2017 10:21

Dear friend, could you more info? Like image or and error code?

Log inor Register

to rate and post a comment

(30)
(9)

Dean Kenney

re: Ct Brain image getting Cut

March 19, 2017 04:59

it's is your ID.10T error which stands for the operator is asking for a tilt of so many degrees for a head scan and then decides when they setup for the scan after they do a scout they see that they are not centered and move the table down or up. the system is not smart enough to know that they moved the table and therefor it cut off the head image..

Log inor Register

to rate and post a comment

(7)

George Anyetei Otchere

re: Ct Brain image getting Cut

March 24, 2017 10:37

In axial scan,25cm is the maximum FOV. So when positioning the head, the lateral line should be at the center of the brain or skull. Now if the scan is cut at the upper part of the brain, then patient position should be moved down and vice versa.

Log inor Register

to rate and post a comment

stats
DOTmed text ad

PHILIPS MX8000 & Brilliance Systems & Spare Parts In Stock Now!

HIGH QUALITY fully tested & refurbished systems to manufacturer specifications like new. Installation & service support worldwide. Parts & tubes in stock with warranty. Sales@Elsmed.com

DOTmed Certified stats

Upcoming Auction Events

Cbrx2sales
TIMED Ending Soon
Jun 05, 2026 (157 lots)