roger dussart
Ct ge hispeed
August 10, 2017 08:09
Hi everybody . Pb calibration hispeed. I cannot calibrate because stop airczl at 140kv. Scan off line resume at 140 kv. I was replaved kv board but the problem is same. Some friend technician can help me to fix this problem? Thanks in advance
to rate and post a comment
Gabriel Cool
re: Ct ge hispeed
August 10, 2017 11:09
Hi. Do you make scans with 140 kv? What is error logs?
CRAIG WAMBURA
August 13, 2017 01:48
Please send me the error log (craig@telemed-technologies.com...I can help. Can you perform offline scans at 140KV, that is with X-ray ON?
Extensive inventory of Philips & Siemens MRI & CT used-tested, refurbished parts. Knowledgeable Sales and Inventory Team. 90-Day Warranty. Bilingual Support. Competitive Pricing. Technical Support. ISO Certified. parts@keimedicalimaging.com
roger dussart
Ct ge hispeed
August 10, 2017 08:09
Hi everybody . Pb calibration hispeed. I cannot calibrate because stop airczl at 140kv. Scan off line resume at 140 kv. I was replaved kv board but the problem is same. Some friend technician can help me to fix this problem? Thanks in advance
to rate and post a comment
Gabriel Cool
re: Ct ge hispeed
August 10, 2017 11:09
Hi. Do you make scans with 140 kv?
What is error logs?
to rate and post a comment
CRAIG WAMBURA
re: Ct ge hispeed
August 13, 2017 01:48
Please send me the error log (craig@telemed-technologies.com...I can help.
Can you perform offline scans at 140KV, that is with X-ray ON?