Sara Canas A

CT GE LIGHT SPEED

May 29, 2018 10:07

Help please, the octane2 of a CT GE Light Speed plus shows the same in both monitors, how do I solve this?

Log inor Register

to rate and post a comment

(423)
(40)
(5)

Juan de la Cruz Martinez

re: CT GE LIGHT SPEED

May 30, 2018 09:36

Reinstall software

Log inor Register

to rate and post a comment

(2)

sharad

re: re: CT GE LIGHT SPEED

May 31, 2018 09:34

i think you are right but this will be last option

Log inor Register

to rate and post a comment

(11)
(5)

Nathan PARRIS

re: CT GE LIGHT SPEED

May 30, 2018 09:36

I did this once. It sound like you loaded the software wrong. That command on the first page tells you what to type if you have two monitors.

Log inor Register

to rate and post a comment

(2)

sharad

re: CT GE LIGHT SPEED

May 30, 2018 09:40

if problem came all of sudden than their is chance someone accidentally press shortcut key for single head selection still u can see horizontal scroll bar to see another monitor on same monitor if problem is same you can try hot keys (along with shift or control) im will also try to find exact key for the same

Log inor Register

to rate and post a comment

(72)
(18)

tim cowell

re: CT GE LIGHT SPEED

May 30, 2018 09:40

The unix command hinv will show the hp hardware that was detected the last time it was booted. Check to see if a dual channel display was detected

Log inor Register

to rate and post a comment