ct ge prospeed

(62)
(10)

Ricardo Rivera

ct ge prospeed

September 03, 2017 12:31

hi every body, I need your help in a ct scanner ge prospeed, when I do warm up apper the error ER 177 XG 102 and 103. the line voltaje is 380 vac in each fase by three . I change all igbts, the rotor control, the xray tube, the HV transformer negative and positive, the configuration of PDU and I have the same error. I am not sure if the dc voltaje of 550 is this voltaje or 700 vdc. because the PDU have 650 vdc aproxx and the other voltajes are down for example the output of 100 I have 90vac.
i change too the PDU.
at this moment I don't know what more can I doo.
if some one have an idea or can help me I appreciate
i am thinking that this ct work only whit 480 vac by fase. thank you

Log inor Register

to rate and post a comment

(181)
(27)

Tibor Ivanszky

re: ct ge prospeed

September 04, 2017 10:26

Hi Ricardo,
Are you sure the error is er-177?
Normally, ER-117 XG errorxxxxxxxxxxx.
Send me the screenshot of the error.
tibor@ivanszky.com

1 person liked this
Log inor Register

to rate and post a comment

(62)
(10)

Ricardo Rivera

re: re: ct ge prospeed

September 05, 2017 10:47

HI Tibor excuse me, the error is ER-117 XG00000103, and XG00000102

1 person liked this
Log inor Register

to rate and post a comment

(138)
(18)
(1)

umberto sciunnach

re: ct ge prospeed

September 04, 2017 10:26

Hello
At first : by where the CT came from ?
USA Europe etc?
Usually is supplied from the customer line 3 phases 380v , the line arrive at a big transformer and the outside is 3 phases 110v.
Let me known.
Regards
Umberto

Log inor Register

to rate and post a comment

(62)
(10)

Ricardo Rivera

re: re: ct ge prospeed

September 05, 2017 10:47

Hi,Umberto, i have 3 lines of 380vac each one and don't have transformer, the three lines of 380 are connecte directly to the PDU

Log inor Register

to rate and post a comment

(62)
(10)

Ricardo Rivera

re: re: ct ge prospeed

September 05, 2017 10:47

excuse me, betwen fase and fase I have 380 vac three lines.

Log inor Register

to rate and post a comment

(62)
(10)

Ricardo Rivera

re: re: ct ge prospeed

September 05, 2017 10:47

Betwen each fase a ground 223 vac

Log inor Register

to rate and post a comment

(410)
(57)

CRAIG WAMBURA

re: ct ge prospeed

September 04, 2017 10:26

Please provide full error code in the format ER 117 XG ERROR 00000XYY....
We need the X and YY values to help you troubleshoot the problem

1 person liked this
Log inor Register

to rate and post a comment

(62)
(10)

Ricardo Rivera

re: re: ct ge prospeed

September 05, 2017 10:47

hi, the full error code ER-117 XG00000103 and XG00000102

Log inor Register

to rate and post a comment

(181)
(27)

Tibor Ivanszky

re: re: re: ct ge prospeed

September 05, 2017 04:25

It is much better to write email to tibor@ivanszky.com.
I have a really good troubleshooting documentation and i can send it to you.
In the meantime i have to say the PDU has many voltage options depending of the origin of the system.
You did not mention that this is a new installation and forum members including me beleived it is a system worked before ok. The errors suggest filament board 15V safe error (102) and CTVRC Fuse blown or overvoltage capacitor failure (103) first clear the voltage problems.

Log inor Register

to rate and post a comment

(205)
(18)
(2)

Pepe Garcia

re: re: re: re: ct ge prospeed

September 07, 2017 10:35

good morning Tibor
as we know in this module is built based on capacitors resistors and resonant circuits, could you tell me a little more about which capacitor of voltage is involved (which would be damaged?).
pepe

Log inor Register

to rate and post a comment

(410)
(57)

CRAIG WAMBURA

re: re: re: ct ge prospeed

September 05, 2017 04:26

ERROR XG00000103:

The X value of 1 indicates that the error occurs during "stand-by"
the YY value of 03 indicates over voltage on the capacitors.
This is a CTVRC Short–circuit Error: A short–circuit (shoot–through) occurred between the time when the generator closed the safety loop and detecting the correct 550 (or 750) VDC in the generator.

ERROR XG00000102:

The X value still indicates error occurs in "stand-by"
the YY value of 02 indicates 15V_Safe Error. The 15_SAFE is used for drivers for the gate control signal.

You have a 15_SAFE Voltage failure. The power supply voltage for optical devices used in the KV Inverter and CTVRC is out of the specified range (13.5 V - 15.5 V). This is detected after all the generator error status is reset and before all the devices are reset.

Check the LED status on the DCRGS Control Board (DS168 - DS171)

Log inor Register

to rate and post a comment

(205)
(18)
(2)

Pepe Garcia

re: re: re: re: ct ge prospeed

September 07, 2017 10:36

as we know in this module is built based on capacitors resistors and resonant circuits, could you tell me a little more about which capacitor of voltage is involved (which would be damaged?).
pepe

Log inor Register

to rate and post a comment

(205)
(18)
(2)

Pepe Garcia

re: re: re: re: ct ge prospeed

September 09, 2017 11:16

hello I have the pdu 2121798 the leds are differents

Log inor Register

to rate and post a comment

(30)
(4)

Martin Martinez

re: ct ge prospeed

September 05, 2017 10:50

Ricardo, good morning.
I can help you, please contact me.
Regards.

Log inor Register

to rate and post a comment

(1)

Siva Prakaash

re: re: ct ge prospeed

October 08, 2017 09:13

Sir, I am having problem in the ge prospeed.. The error code which I got is error 117 XG00000211 . what is the solution for this

Log inor Register

to rate and post a comment

(205)
(18)
(2)

Pepe Garcia

re: re: re: ct ge prospeed

October 09, 2017 08:17

check these points..........
117 201,211,311 bad contact connector bad contact on XJ13 E-rack
117 201,211 Slip ring slip ring very dirty + pb fonct tilt. Reajusted +5v OBC (4.93v) ER103 ??
117 211 intermiitent error large PDU FE changed PDU CTRL board and problem was solved
117 201/211 with big noise in gantry CTVRC Cable on ctvrc was burned
117 201,211,218 optical fiber MR kits were used (part no: 46-301450G1 and 46-317830G1)
117 201 SRI mini pdu PDU ctrl board bad relay k4 which command relay k1 ERR 117/211
117 201 intermittent OPT cable E-rack CTVRC OPT cable ER117/211/218

Log inor Register

to rate and post a comment

(205)
(18)
(2)

Pepe Garcia

re: re: re: ct ge prospeed

October 09, 2017 08:17

please email me

Log inor Register

to rate and post a comment