Page 329 of the service manual. This is a rotor failure. Check the rotor controller and the stator circuit at the tube. It is generated by a timeout from the rotor ready signal.
GCI offers complete MRI & CT sales & service with ROI analysis, in-house financing, contractor drawings, accreditation, training, marketing, and OEM-level support beyond the sale.
simon ramirez
CT MAX OF GENERAL ELECTRIC, ERROR CODE
November 19, 2006 04:33
THE CODE ERROR IS ER-123 XG FAIL THIS ERROR IS FROM TUBE? OR WHAT THANKS
to rate and post a comment
Ralph Childs
Re: CT MAX OF GENERAL ELECTRIC, ERROR CODE
November 20, 2006 12:30
Simon,
Page 329 of the service manual. This is a rotor failure. Check the rotor controller and the stator circuit at the tube. It is generated by a timeout from the rotor ready signal.
Scooter
scanner1@eim.ae
to rate and post a comment
Abel Cardona
Re: CT MAX OF GENERAL ELECTRIC, ERROR CODE
November 21, 2006 05:44
check the rotor system
Im in mxico, let me know if can i help you.
regards.
Abel
to rate and post a comment
Marcelino Carmona
Re: CT MAX OF GENERAL ELECTRIC, ERROR CODE
November 21, 2006 10:37
This is a rotor controller error and you can check that manually let me see if i can help you
good look
Carmona