Over 150 Missouri Auctions End Tomorrow 06/05 - Bid Now

CT Pace problem

(29)
(5)

Mohamad Ismail

CT Pace problem

November 10, 2005 02:54

Dear All ,
I have a problem in a CT GE Pace, the message " X-Ray Zero Detect" appears when trying to warmup,there is a X-Ray for 4-5 seconds of warmup. I checked the 1000V power supply - OK,
Can anybody help ?

Log inor Register

to rate and post a comment

Horia Ioan

Re: CT Pace problem

November 10, 2005 09:44

Hello
Go in service mode, support display and follow procedures to display a profile of x-ray beam

good luck
Horia Ioan

Log inor Register

to rate and post a comment

(47)

Mike Lynch

Re: CT Pace problem

November 10, 2005 11:26

You most likely have a bad channel on one of your cam boards. Do a tube warm up series and look into the support display program and set up for HPF. This will plot your offsets and tell you what channel, therefore what cam board, is dead.

Log inor Register

to rate and post a comment

(53)
(2)
(1)

yasar demirel

Re: CT Pace problem

November 10, 2005 04:22

I had same problem last year
when you measure 1000 v after disconnecting from dedector you may not be sure
pls check voltage from dedector
ydemirel@istanbulmedikal.com

Log inor Register

to rate and post a comment

(29)
(5)

Mohamad Ismail

Re: CT Pace problem

November 16, 2005 02:33

Dear Sir

Thank you for your support.
I turned on the system to followup with your procedure but the system working with no errors, it seems that the problem is intermittent.

Regards, Mohamad.

Log inor Register

to rate and post a comment

(120)
(6)
(1)

manuel valdez

re: Re: CT Pace problem

April 19, 2013 09:17

hi Mike i have a problem with a sytec 2000i each time that i do a warmup the system stop and show the error 122 i tried view channels damaged with the support display but when i write the file name SUPRAW1 the console show *FILE NOTHING i test the power supply and is ok
the last study in the disk show many artifacts diagonal and cruize lines please help me

Log inor Register

to rate and post a comment

(47)

Mike Lynch

Re: CT Pace problem

November 16, 2005 02:28

If the problem went away, you either had corrupt calibration files, which happens regularly with YMS, or you have an intermittent cam board and/or DAS power supply. Also, check the power supply on the OBC, I have seen this introduce intermittent errors.

Log inor Register

to rate and post a comment

stats
DOTmed text ad

MRI Solutions to Meet Your Team’s Unique Needs

From preventive maintenance through comprehensive cryogen services, BMI-PMI is your trusted partner for new/reconditioned sales, installation, and service of both facility-based and mobile MRI solutions. www.brownsmedicalimaging.com / 800-701-9729

stats

Upcoming Auction Events

Cbrx2sales
TIMED Ending Soon
Jun 05, 2026 (157 lots)