1-on the checkup window when it reach test image and after we cofirm gantry
rotation and request for start once we hit the start buttom the reset
aborted , the massege on the log book says
we notice that we can do x ray on gantry stationry
but during gantry rotation the reset is immediatly aborted
also i would like to inform you that we reinstalled the software VC10C and
restore the previos tables
2- 2nd problem is that when we click on error massege the 2nd window for
remedy&(what to do ) does not appear although we had already intered the
service password.
DirectMed Imaging specializes in Equipment Sales, Parts, and Support. Parts are in stock and available for immediate delivery. In-room MRI Upgrades are available now. (855) 463-3727 or sales@directmedimaging.com
Abdulkarim Sadek
CT Plus 4 Power Problems
October 17, 2006 10:20
1-on the checkup window when it reach test image and after we cofirm gantry
rotation and request for start once we hit the start buttom the reset
aborted , the massege on the log book says
MAS_Ea00030 Cal.table_missg ('E'a0 0 30 1 8 15 0 )
we notice that we can do x ray on gantry stationry
but during gantry rotation the reset is immediatly aborted
also i would like to inform you that we reinstalled the software VC10C and
restore the previos tables
2- 2nd problem is that when we click on error massege the 2nd window for
remedy&(what to do ) does not appear although we had already intered the
service password.
to rate and post a comment
gustavo fiochi
Re: CT Plus 4 Power Problems
October 17, 2006 12:00
you need the extended service password to get more info for the error
to rate and post a comment
Abdulkarim Sadek
Re: CT Plus 4 Power Problems
October 20, 2006 04:20
Thanks, The password is veiled till August 2007.
to rate and post a comment