Abdulkarim Sadek

CT Plus 4 Power Problems

October 17, 2006 10:20

1-on the checkup window when it reach test image and after we cofirm gantry
rotation and request for start once we hit the start buttom the reset
aborted , the massege on the log book says

MAS_Ea00030 Cal.table_missg ('E'a0 0 30 1 8 15 0 )

we notice that we can do x ray on gantry stationry

but during gantry rotation the reset is immediatly aborted

also i would like to inform you that we reinstalled the software VC10C and
restore the previos tables

2- 2nd problem is that when we click on error massege the 2nd window for
remedy&(what to do ) does not appear although we had already intered the
service password.

Log inor Register

to rate and post a comment

(55)
(1)
(1)

gustavo fiochi

Re: CT Plus 4 Power Problems

October 17, 2006 12:00

you need the extended service password to get more info for the error

Log inor Register

to rate and post a comment

(3)

Abdulkarim Sadek

Re: CT Plus 4 Power Problems

October 20, 2006 04:20

Thanks, The password is veiled till August 2007.