Over 150 Missouri Auctions End Tomorrow 06/05 - Bid Now

CT Prospeed error 101;53 error 117;441

Tags: CT, GE

(35)
(3)

Jose Eduardo Martinez

CT Prospeed error 101;53 error 117;441

May 02, 2014 03:32

Good morning I´ve trouble with Prospeed CT the tube has a big rotor´s noise the machine report:
ER-101 Scan error LCP=03TNseqr Adress00245E62
00000053
ER-117 XG ERROR LCP=03TNseqr Adress00245E62
00000441
Some body could help me to resolve this problem?
eduardo

Log inor Register

to rate and post a comment

(317)
(22)
(8)

Joseph Jenkins

re: CT Prospeed error 101;53 error 117;441

May 03, 2014 01:55

If I remember correctly the 117 is a tube fault error ( time to replace?)
The 101 is a general error... The 117 is the key.

Log inor Register

to rate and post a comment

(140)
(35)

FranciscoG

re: re: CT Prospeed error 101;53 error 117;441

October 02, 2016 02:52

Hi Joseph, I'm getting the ER-101 Scan Error LCP=03TNseqr Adress00245E62
00000053

I have another good parts system so I have replaced all the boards from the OGP module, console and TGP, and still getting this error. I cleaned up the complete slipring as well, finding scratches on it.

Also when I press the SET button at the console, there is a led turning ON on the OGP board, the D2A ERRS led. it goes off everytime I reset the board manually or reset the complete system.

If you have any other idea what to do, please let me know.

Log inor Register

to rate and post a comment

(73)
(8)

Kim Schofield

re: CT Prospeed error 101;53 error 117;441

May 03, 2014 01:56

Change the Xray tube

Log inor Register

to rate and post a comment

(34)
(5)
(4)

Mario Icherenski

re: CT Prospeed error 101;53 error 117;441

May 03, 2014 01:56

Dear Jose,
You should replace your X-ray tube MX 165. Probably when you attempt to exposure anode would tear at acceleration and will flow much oil after breaking the X-ray tube.

Log inor Register

to rate and post a comment

(55)
(6)
(2)

Charbel Tadros

re: CT Prospeed error 101;53 error 117;441

May 03, 2014 01:56

Tube problem . Change the tube

Log inor Register

to rate and post a comment

(26)
(2)

Cheema

re: CT Prospeed error 101;53 error 117;441

May 06, 2014 11:52

Hi Jose,

The ER- 117 is telling that you have problem in High voltage section. While Code 00000441 is telling that you have problem at your Anode(+) side.
Please take out the ANODE candels from bothside( Tank and tube. clean and put new HV oil.
If not you have to replace.
i) HV Cable
ii) Anode Invertor.
iii) Anode Tank.
Then Tube.

We have all Parts of Prodpeed Solidstate Detector and Xenon CT scan.

Please Contact at Cheemat@yahoo.com

Log inor Register

to rate and post a comment

(34)
(5)

SEASBI

re: re: CT Prospeed error 101;53 error 117;441

June 26, 2014 04:33

hi good morrning I need support Can you mailme?

Log inor Register

to rate and post a comment

(26)
(2)

Cheema

re: CT Prospeed error 101;53 error 117;441

May 06, 2014 11:52

Please do not suspect your tube first you have to diagnose.

Log inor Register

to rate and post a comment

(317)
(22)
(8)

Joseph Jenkins

re: CT Prospeed error 101;53 error 117;441

October 04, 2016 08:48

your tube is gone....117

Log inor Register

to rate and post a comment

stats
DOTmed text ad

PHILIPS MX8000 & Brilliance Systems & Spare Parts In Stock Now!

HIGH QUALITY fully tested & refurbished systems to manufacturer specifications like new. Installation & service support worldwide. Parts & tubes in stock with warranty. Sales@Elsmed.com

DOTmed Certified stats

Upcoming Auction Events

Cbrx2sales
TIMED Ending Soon
Jun 05, 2026 (157 lots)