Good morning I´ve trouble with Prospeed CT the tube has a big rotor´s noise the machine report:
ER-101 Scan error LCP=03TNseqr Adress00245E62
00000053
ER-117 XG ERROR LCP=03TNseqr Adress00245E62
00000441
Some body could help me to resolve this problem?
eduardo
Hi Joseph, I'm getting the ER-101 Scan Error LCP=03TNseqr Adress00245E62
00000053
I have another good parts system so I have replaced all the boards from the OGP module, console and TGP, and still getting this error. I cleaned up the complete slipring as well, finding scratches on it.
Also when I press the SET button at the console, there is a led turning ON on the OGP board, the D2A ERRS led. it goes off everytime I reset the board manually or reset the complete system.
If you have any other idea what to do, please let me know.
Dear Jose,
You should replace your X-ray tube MX 165. Probably when you attempt to exposure anode would tear at acceleration and will flow much oil after breaking the X-ray tube.
The ER- 117 is telling that you have problem in High voltage section. While Code 00000441 is telling that you have problem at your Anode(+) side.
Please take out the ANODE candels from bothside( Tank and tube. clean and put new HV oil.
If not you have to replace.
i) HV Cable
ii) Anode Invertor.
iii) Anode Tank.
Then Tube.
We have all Parts of Prodpeed Solidstate Detector and Xenon CT scan.
HIGH QUALITY fully tested & refurbished systems to manufacturer specifications like new. Installation & service support worldwide. Parts & tubes in stock with warranty. Sales@Elsmed.com
Jose Eduardo Martinez
CT Prospeed error 101;53 error 117;441
May 02, 2014 03:32
Good morning I´ve trouble with Prospeed CT the tube has a big rotor´s noise the machine report:
ER-101 Scan error LCP=03TNseqr Adress00245E62
00000053
ER-117 XG ERROR LCP=03TNseqr Adress00245E62
00000441
Some body could help me to resolve this problem?
eduardo
to rate and post a comment
Joseph Jenkins
re: CT Prospeed error 101;53 error 117;441
May 03, 2014 01:55
If I remember correctly the 117 is a tube fault error ( time to replace?)
The 101 is a general error... The 117 is the key.
to rate and post a comment
FranciscoG
re: re: CT Prospeed error 101;53 error 117;441
October 02, 2016 02:52
Hi Joseph, I'm getting the ER-101 Scan Error LCP=03TNseqr Adress00245E62
00000053
I have another good parts system so I have replaced all the boards from the OGP module, console and TGP, and still getting this error. I cleaned up the complete slipring as well, finding scratches on it.
Also when I press the SET button at the console, there is a led turning ON on the OGP board, the D2A ERRS led. it goes off everytime I reset the board manually or reset the complete system.
If you have any other idea what to do, please let me know.
to rate and post a comment
Kim Schofield
re: CT Prospeed error 101;53 error 117;441
May 03, 2014 01:56
Change the Xray tube
to rate and post a comment
Mario Icherenski
re: CT Prospeed error 101;53 error 117;441
May 03, 2014 01:56
Dear Jose,
You should replace your X-ray tube MX 165. Probably when you attempt to exposure anode would tear at acceleration and will flow much oil after breaking the X-ray tube.
to rate and post a comment
Charbel Tadros
re: CT Prospeed error 101;53 error 117;441
May 03, 2014 01:56
Tube problem . Change the tube
to rate and post a comment
Cheema
re: CT Prospeed error 101;53 error 117;441
May 06, 2014 11:52
Hi Jose,
The ER- 117 is telling that you have problem in High voltage section. While Code 00000441 is telling that you have problem at your Anode(+) side.
Please take out the ANODE candels from bothside( Tank and tube. clean and put new HV oil.
If not you have to replace.
i) HV Cable
ii) Anode Invertor.
iii) Anode Tank.
Then Tube.
We have all Parts of Prodpeed Solidstate Detector and Xenon CT scan.
Please Contact at Cheemat@yahoo.com
to rate and post a comment
SEASBI
re: re: CT Prospeed error 101;53 error 117;441
June 26, 2014 04:33
hi good morrning I need support Can you mailme?
to rate and post a comment
Cheema
re: CT Prospeed error 101;53 error 117;441
May 06, 2014 11:52
Please do not suspect your tube first you have to diagnose.
to rate and post a comment
Joseph Jenkins
re: CT Prospeed error 101;53 error 117;441
October 04, 2016 08:48
your tube is gone....117
to rate and post a comment