CT scanner room shielding

Tags: CT, PHILIPS

(16)
(2)
(2)

Antonio Lopez

CT scanner room shielding

October 30, 2023 10:31

Hello, can someone tell me what is the thickness of lead used in shielding the room of a 128-slice CT scanner? The room dimensions are 14.6 feet (4.45m) x 23.6 feet (7.18m). Thank you

Log inor Register

to rate and post a comment

(1)

BKindia

re: CT scanner room shielding

October 31, 2023 10:37

In general, a lead shielding of 2mm thickness should suffice.
However, it is recommended to check and comply with your local regulatory authorities if they have any specific guidelines regarding this.

Log inor Register

to rate and post a comment

(23)
(4)

Mohammed Boudria

re: CT scanner room shielding

November 01, 2023 01:14

2 mm eq Pb , coukd be less, depending on thickness and materials of the walls.
Height should be 2 m or more depending of the rooms/areas floor levels level with respect to scan room.
Take into account the radiation curves of the ct.